nice design!
 
Here are some thoughts from my side:
 
The LT1208 has a fairly significant Voffset-versus-temp sensitivity. Maybe  a 
part like the LTC6244HV with about 1/10 the offset drift would work better,  
and reduce the thermal sensitivity of the overall system.
 
For example, the LT1208 has 25uV per Degree C drift, and with a standard  MTI 
OCXO which has about 40Hz deviation from 0V to 5V EFC this would be about  
2E-011 drift per degree Celsius worst case just due to the opamp drift, a fairly 
 significant amount.
 
The LTC6244HV part only allows a power supply range of +-5V so you may  want 
to add low-noise 5V and -5V regulators for the opamps.
 
 
The feedback and reference resistors R1, R2, R5, R6 are fairly high  value, 
and thus will add noise to the EFC. Make sure to use metal film types  here. To 
reduce overall noise, you may want to add a simple RC lowpass filter to  the 
output of the opamp, say four X7R 22uF10V caps driven by a 4.7K  metal film 
resistor etc.
 
Ground loop current coming from the OCXO heater through the OCXO ground pin  
and the ground plane can also cause some EFC error. Make sure to route the  
OCXO ground to the OCXO ground pin on the Fury, and then separate-out the EFC  
ground through a star-type layout. The idea is to prevent the OCXO heater 
ground  current from flowing through the EFC circuitry and SMA cable, in which it 
would  cause voltage drop and thus thermal sensitivity.
 
Also, you may want to allow the OCXO to be powered from the Fury OCXO  
footprints' power pins. The Fury can compensate for thermal sensitivity if this  is 
done. Alternatively, a thermistor may be connected to the OCXO, and driven by  
the Fury's OCXO power pins (thermistor must handle 10.5V!). As long as the  
current varies with temperature on the OCXO power pins, the Fury will learn, 
and  compensate thermal sensitivities in the system. BTW: the thermal tempco  
measurement is disabled via software command by default on the SMA Fury's, and  
can be enabled via SCPI command.
 