wje wrote:
> In this case, the temp thermistor bridge is outside the oven cavity
> itself. The cable only passes power and the already-processed bridge
> delta to the heater power amp. So, there's no particular benefit from
> having the cable stuck to the heater wrap. (at least, I think so; my
> basic failure was because the cable fried and shorted power to ground)
> Bill Ezell
> ----------
>
Bill
If the temperature bridge is outside the oven cavity then its critical
that the temperature sensor leads are thermally shunted to the oven.
Substituting constantan wire for copper wire also helps as the thermal
conductivity of constantant is significantly lower than that of copper.
For example if the temperature sensor has a thermal resistance of 1K/W
to the oven and the leads have a thermal resistance of 100K/W to ambient
then ambient temperature fluctuations of 10 K will induce temperature
sensor temperature variations of 0.1K which the oven controller will
correct by varying the oven temperature by 0.1K.
The lead themal resistance would have to be > 1E4K/W to maintain oven
temperature fluctuations below 1mK when the ambient temperature varies
by 10K.
Such a high thermal resistance is difficult to achieve.
The thermal resistance of a length of wire can be estimated by measuring
its electrical resistance and dividing it by the product of the thermal
conductivity and electrical resistivity.
For copper wire thermal resistance ~ 1.6E5 x Electrical resistance
For 10 cm of 20swg Cu wire the electrical resistance (at 20C) is about
4.4 milliohms and the corresponding thermal resistance is about 700 W/K.
Bruce