[time-nuts] D flip flop 1 bit phase detector
Bruce Griffiths
bruce.griffiths at xtra.co.nz
Sun Sep 21 18:22:39 EDT 2008
The D flipflop phase detector where the clock input is driven by the PPS
output of a GPS timing receiver, the D input is driven by the OCXO
frequency (or a subharmonic thereof), and the filtered Q output is
sampled after a sufficient delay to reduce the probability of a
metastable output to an insignificant level and used (after filtering)
to drive the EFC input of the OCXO to be disciplined can be viewed as
akin to a sigma delta loop operating in the phase domain. The intrinsic
noise and or quasi sawtooth jitter of the PPS output serve to break up
idle tones that may otherwise be present.
For correct operation the period of the D input frequency should be
greater than the PPS sawtooth dither or intrinsic PPS jitter.
The loop will inherently lock the frequency at the D flipflop input to a
harmonic of 1Hz, so to ensure the OCXO locks to the desired frequency
either the tuning range must be restricted or the OCXO frequency be
divided down to a suitable frequency that precludes the OCXO locking to
the wrong harmonic.
If hardware sawtooth correction is used on an M12M timing receiver the
residual PPS jitter drops to a few nanosec and the effective resolution
of the D flipflop phase detector adapts to a fraction of that.
Bruce
More information about the time-nuts
mailing list