John

One can cure the propagation delay tempco associated with a 74HC390
divider string by resynchronising the output to the input clock.
However worst case design means that a 3 stage synchroniser is required.
Assuming a 10Mhz input clock frequency:
First resychronise the output to 1MHz (worst case propagation delay for
PPS output is < 1us)
Then resynchronise this output to 2MHz (assumes a biquinary
configuration for the 1st 1/2 of the input '390)
Finally resynchronise this output to the 10MHz input clock.

Bruce

John Ackermann N8UR wrote:
> Hi Brian --
>
> It's good to collect this data; thanks.  It's interesting that your std
> dev in the first test seems to increase significantly with the number of
> samples; I haven't seen that kind of scaling here (1K sample and 10k
> sample turned in very similar std dev).  From what Poul-Henning said
> earlier, your first run may suffer the same distortion as my data at the
> bottom of this thread.
>
> I just finished rerunning the TADD-2 test using a Wavecrest DTS-2075
> (the first real use I've had for that box!) and with 1 PPS input on the
> start channel, 10 MHz from the same source on the stop channel, and 10K
> samples, I got 22.0 ps of jitter, and a 92 ps min/max range.  (As far as
> I can determine, the Wavecrest doesn't allow you to use an external
> reference, and its internal reference runs at 100 MHz so it probably
> wouldn't be useful in this measurement.)
>
> That's consistent with what I measured earlier with the 5370B when I
> didn't have the reference and the inputs in coherence.  It appears that
> the test below, where I used the same reference for *everything*
> triggered the problem that Poul-Henning warned about, so those results
> should be disregarded.
>
> While I haven't done any testing to validate this, I think the complaint
> about the 74HC390 dividers isn't so much their jitter in normal use, but
> the tempco problems the cascaded stages can cause.  If you can do it, it
> would be interesting to measure the phase change over temperature --
> I've done a preliminary experiment on that for the TADD-2, but plan to
> rerun it with much better measurement technique.
>
> I'm also hoping to do a jitter and tempco test of the Wenzel input
> conditioning circuit by itself.  I really like that circuit for its wide
> input amplitude range.
>
> John
> ----
> Brian Kirby said the following on 04/04/2009 04:18 PM:
>   
>> I will report some results on a asynchronous divider, which I basically 
>> copied from Dr. Thomas Clark's designs, which everybody likes to report 
>> as a bad design.
>>
>> The 10 MHz input signal is coupled thru a resistor and capacitor.  On 
>> the other side of the capacitor is the resistive divider that is tied to 
>> Vcc and ground - it biases the signal to 2.5 volts, which is feed to the 
>> input of the 74HC132.   The output of the 74HC132 feeds several 74HC390s 
>> until it becomes a buffered 1 pulse per second signal.  I also have 
>> buffered 5 MHz and 1 MHz outputs.  The other 3/4 of the 74HC132 are used 
>> to externally synchronize the 74HC390s.
>>
>> I used the Thunderbolt as the source of 10 MHz and it was feed to the 
>> divider, and the stop input on the HP5370B.  The 5370B was run on 
>> internal clock.  The 1 PPS from the divider feed the start input on the 
>> 5370B.
>>
>> 100 seconds   TI 79.865 nS   MIN 79.80 nS   MAX 79.98 nS   STD 36.4 pS.
>> 1000 seconds   TI 79.831 nS   MIN 79.71 nS   MAX 80.00 nS   STD 49.9 pS
>> 10K seconds   TI   80.1552 nS   MIN 79.79 nS MAX 80.88 nS   STD 271 pS
>> 100K planned
>>
>> Also a second test, using the Thunderbolt as a source of 10 MHz and it 
>> was  feed to the divider, the stop input on the 5370B and the external 
>> clock of the 5370B.  The 1 PPS from the divider feed the start input on 
>> the 5370B.
>>
>> 100 seconds   TI   75.002 nS   MIN 74.96 nS   MAX 75.04 nS   STD 22.5 pS
>> 1000 seconds   TI    74.931 nS   MIN 74.80 nS  MAX 75.04 nS   STD 56.8 pS
>> 10K seconds   TI   77.5135 nS  MIN 77.40 nS  MAX 77.62 nS  STD 35.9 pS
>> 100K measurement in progress.
>>
>> I believe having STD in parts of 10-14th is fairly respectable for 
>> amateur designs..
>>
>> Brian KD4FM
>>
>> John Ackermann N8UR wrote:
>>     
>>> I just finished a jitter test of the first TADD-2 built on the 
>>> production circuit board.
>>>
>>> The configuration was somewhat optimized from what I used for the 
>>> earlier tests.
>>>
>>> A single 10 MHz source was daisy-chained to the TADD-2 input, to the 
>>> 5370B external reference input, and to the 5370B STOP channel.  The 1 
>>> PPS output from the TADD-2 was connected to the 5370B START channel. 
>>> Thus any reference jitter shouldn't be common-mode, and using the 
>>> reference clock on the STOP channel avoids the need for a second 
>>> divider, and ensures that the time interval is small (always less than 
>>> 100 ns; in this case, about 90 ns).
>>>
>>> For a 10,000 sample run, the standard deviation was 12.1 picoseconds, 
>>> and the peak-to-peak variation was 70 picoseconds.  Based on experiments 
>>> I ran a few years ago, I think this is pretty much the noise floor of 
>>> the 5370B and the divider could be better than this.
>>>
>>> John
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>>   
>>>       
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>     
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>   