Esa Heikkinen wrote:
>> The capacitance of the 1x probe is too high.
>>
>
> So I have to check that also: it shows 46 pF on 1x mode and 10.6 pF in
> 10x mode.
>
> Ok so the main question is that what kind of amplifier I should use with
> 10544A to get a clean signal, if it can't stand even an oscilloscope
> without distortion on it's output? And what do that kind of amplifier
> cost - more than HP10811, for example?
>
> Many other OCXO's (like thunderbolt's) can be connected directly to the
> 50 ohms. load without any trouble with harmonics.
>
>
Esa
You can easily construct a suitable buffer amplifier using inexpensive
and readily available parts.
It was an early design, intended to drive a 1k load located very close
to the connector terminals.
The buffer design depends on what reverse isolation you require and what
output signal level.
The venerable LH0033C has been used as a buffer to drive a 50 ohm load.
However it is relatively expensive if its even available.
Since this is RF you dont need the low offset voltage and drift of such
a device.
Bruce