[time-nuts] fleabay cesium
WB6BNQ
wb6bnq at cox.net
Sat May 9 09:23:40 UTC 2009
Magnus,
That is why I said ". . . setup and properly adjusted . . "
Bill....WB6BNQ
Magnus Danielson wrote:
> WB6BNQ skrev:
> > Well Dick,
> >
> > Assuming it is setup and properly adjusted, it is absolute as a
> > frequency reference by "definition." It does have a given resolution
> > as stated in the manufacturer's manual for the instrument. This
> > resolution [or noise floor as I see it] is due to a number of factors,
> > not least of which is the condition of the physics package (the actual
> > Cesium tube). They do ware out !
>
> Actually, Cesiums can both be offset in frequency and have drift
> components in them. There are a number of systematic issues to handle.
> The magnitude of these depends on how they are handled in the design as
> well as how they are maintained.
>
> Also, general relativity comes in so the systematic error of running the
> clocks above the gravity of standard sea level acceleration comes into
> mind. For a relative comparision at that site it does not care that much
> usually.
>
> There are many features of a Cesium beam which affects the precision,
> but none of them actually runs at the magical number, they are all
> skewed by the C-field and compensation for this needs to be done.
> However, while the C-field skew can be calculated, it is hard to know
> the actual C-field. Modern "digital" Cesiums use a mechanism to measure
> and stabilize the C-field by measuring the position of the nearby Rabi
> responses in addition to the central Rabi and Ramsey features. Such an
> approach can remove the major part of the frequency error traceable to
> that systematic issue. For older cesium beams you adjusted the current
> in the C-field to match up, so calibration to match frequency is needed.
>
> Older Cesiums does not have this "digital" control loop, so it would be
> inaccurate to say that they by definition is sharp on. They are not and
> never really be sharp on.
>
> There is a number of other stability issues of various forms to add onto
> this, but I think it is out of scope here.
>
> Cheers,
> Magnus
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the time-nuts
mailing list