[time-nuts] Homebrew Maser
Corby Dawson
cdelect at juno.com
Mon Aug 30 16:09:39 UTC 2010
John,
>Initial tuning is not the bugaboo. It's fairly easy to set the frequency
with a synthesizer as a probe into the cavity via the circulator.
Once within 10Khz or so you then pick up the difference with the cavity
temp.
You adjust so that the cavity is set within the varactor tuning range.
Remember the cavity is regulated to within .001 degree C to help
stabilize the frequency.
>The external insulating cavity wall is exposed to the differential but
the internal tuned cavity has a vacuum inside and out so there is no
differential pressure on it.
>This Maser does not have cavity auto tune so is a bit less complex in
that regard. Spin exchange tuning is the best way to get to the proper
operating point.
this requires some measurements and then setting the cavity varactor to
the calculated voltage.
Corby Dawson
____________________________________________________________
Obama Urges Homeowners to Refinance
If you owe under $729k you probably qualify for Obama's Refi Program
http://thirdpartyoffers.juno.com/TGL3141/4c7bd7f763237d4c80m04duc
More information about the time-nuts
mailing list