[time-nuts] Frequency comparison
Bob Camp
lists at cq.nu
Tue Feb 9 12:38:39 UTC 2010
Hi
Efratom became part of Datum a while ago. They changed the marking on the LPRO's after the ownership change.
Same product, just a new name on the door.
Bob
On Feb 9, 2010, at 7:28 AM, Joop wrote:
>> It is best to leave your stable quartz and Rubidium oscillators
>> running continuously, preferably on a UPS in case of power failure,
>> for best results.
>>
>> Best regards,
>>
>> Charles
>>
>
> I can understand. But I will not need that high accuracy very often.
> So I just wanted to examine what can be expected after power-on. Just
> saving a bit of environment where possible.
>
>> From a Datum LPRO specifications document I read:
>
> Warm-up: (at -20°C) (at 25°C)
> Time to Lock: < 8.7 min < 5.4 min
> Time to <1E-9: <10.2 min <7.3 min
> Time to <4E-10: <12.7 min <10.6 min
>
> Frequency Retrace: <±2.5E-11
> (after 24 hrs power on @ 25°C & up to 48 hrs power off)
>
> I guess my readings are all within the above specs. The retrace value is
> probably the limiting factor. 2.5E-11 translates in my setup to 90ns/hour.
> Which looks being reached already in about 2 hours after power-on. So if
> the Efratom is essentially the same as the Datum, then all looks fine.
>
> Joop
>
>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>