[time-nuts] DMTD to MMTD
Bob Camp
lists at cq.nu
Fri Feb 19 03:02:00 UTC 2010
Hi
Anything over 100 MHz should do the trick. On a machine with a full OS on it that kind of sustained I/O would be tough to do un-interupted.
Now, doing a 32 bit shift register externally and clocking in a word at 100 / 32 MHz - not quite as hard.
Bob
On Feb 18, 2010, at 9:54 PM, Bruce Griffiths wrote:
> Magnus Danielson wrote:
>> Bruce Griffiths wrote:
>>> Magnus Danielson wrote:
>>>> Bruce Griffiths wrote:
>>>>> Lux, Jim (337C) wrote:
>>>>>>> -----Original Message-----
>>>>>>> From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On Behalf Of Bruce Griffiths
>>>>>>> Sent: Wednesday, February 17, 2010 2:10 PM
>>>>>>> To: Discussion of precise time and frequency measurement
>>>>>>> Subject: Re: [time-nuts] DMTD to MMTD
>>>>>>>
>>>>>>> The latest version actually records time stamps from a continuously
>>>>>>> running counter clocked at some at a constant frequency (100Mhz??)for
>>>>>>> all channels simultaneously.
>>>>>>> They may use a flag bit for each for each channel to indicate to which
>>>>>>> channel or channels the zero crossing time stamp belongs.
>>>>>> Simpler than that.. it grabs 20 bit numbers and shoves them out in ASCII over a com port with an indication of which channel it was for.
>>>>>> The FPGA has a 20 bit free running counter at 100 MHz. When an input changes state, it latches the counter, and shoves it out along with the channel number. They use an offset frequency>100 Hz so that you don't have to disambiguate the counter rollovers. (20 bits rolls over every 10+milliseconds counting at 100 MHz)
>>>>>>
>>>>>> I don't know if there's a FIFO in front of the UART (e.g. what if you get simultaneous zero crossings).. but I would expect there is.
>>>>>>
>>>>>> The "hard work" is in the zero crossing detector ahead of the FPGA. (and perhaps in the latching of the ZCD inputs into the FPGA).
>>>>>>
>>>>>> Given how long ago it was made, that FPGA isn't a huge one.
>>>>>>
>>>>> Using 8 flag bits (one per channel) together with the associated time stamp is a little more efficient and very easy to do and it doesn't require a FIFO to ensure that simultaneous zero crossings aren't missed.
>>>>
>>>> It doesn't help at all for this application. The 8 channels is more likely to be spread out and as Jim pointed out, the next bin is sufficient for needing a unique time-stamp. The flag-solution is less efficient (8 bits rather than 3) and the FIFO need is always there, but may be implemented in various ways. For the flag system to be efficient, a high probability for the same time-bins to be used needs to exist, and a high resolution system can expect to actually see noise spread out the channels over the time-bins.
>>>>
>>> Depending on the system constraints it may be the difference between being able to do implement it or not.
>>
>> With your clarifying comment yes... because it is until now that you have clarified the merit of the flag system, lowered implementation complexity vs. lowered signalling capacity.
>>
>>>> A DMTD systems have a low rate of events per channel, but the nominal distance of events for each channel is fairly long, worst-case burst is when all channels time-stamps. For a 100 Hz beating and 100 MHz clock, the nominal rate of rise/fall events is 200 Hz or 5 ms. Letting the locked value stay put for at least 4 ms. If all channels could be emptied within these 4 ms (just another way of saying that it has enough transport capacity) then a fairly simple schedule system can loop through the channels to find a new sample to transmit.
>>>>
>>>> I think the re-occurring flag system should be put to rest, it doesn't contribute and is a red herring, at least for this application.
>>>>
>>> Nonsense, it requires simpler logic and for a device with limited internal connection/routing capability and a large number of channels the data path interconnections may be simpler and easier to route. It may also run with a higher clock frequency.
>>>
>>> It should even be possible to impement in a relatively small CPLD albeit with an external FIFO or equivalent (eg a PPI port on a Blackfin DSP).
>>> Each additional channels requires one input pin, one output pin, a 2 bit synchroniser and a 1 bit wider data path and little else.
>>
>> But it produces more data, which was what I was commenting on. Indeed it is very simple to implement, but it's a complexity which is still on the low-end.
>>
>> So, finally you made the point of the merit of the approach in such a way that it became clear to me why you have maintained that standpoint.
>>
>> Cheers,
>> Magnus
>>
>>
> One can always reduce the data sent to the PC by removing redundant information, or increase it by Ascii coding of channels and slope polarities.
> In general in the absence of significant constraints either method will work, they just have different tradeoffs.
>
> Its tempting to see if one can just synchronously clock the synchronised ZCD output signals directly into a processor and derive the timestamps entirely in software.
> The only question being what is the maximum clock frequency at which this can be done without missing any ZCD transitions.
> A Blackfin DSP for example should be able to do this with a maximum ZCD sampling clock frequency of at least 50MHz.
>
> Bruce
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the time-nuts
mailing list