[time-nuts] No State Of The Art Counter
Hal Murray
hmurray at megapathdsl.net
Fri Jan 7 06:27:25 UTC 2011
[context is SERDES inputs to FPGA]
> Also, one has to count the number of 0->1 transitions in each word, such
> that the event counter can be accumulated with the contribution from each
> word.
That gets complicated. My straw man would be to get a single transition per
word working first.
Let's check the numbers.
If the serial bit rate is 3.2 gigabits, that's 100 MHz per 32 bit word. If
you are limited to 1 rising edge per word, that limits your counting to 100
MHz.
If you are interest in ADEV of a 10 MHz source, 100 MHz is a fine limit.
Below, I'm assuming "transition" means rising edge.
So my first proposal for an API between the front end and the back end would
be:
5 bits of offset for the transition
1 bit for "Yes, there was a transition"
1 bit for error (More than one transition)
Plan two would be something like:
5 bits of offset for last transition.
n bits for number of transitions
(with 0 for none, and xx for too many, aka error)
--
These are my opinions, not necessarily my employer's. I hate spam.
More information about the time-nuts
mailing list