What they do does not make any difference to us, what is important how
often we have to change the tuning word to meet our requirements, and if we
dither it will be an addition or subtraction from the center value. With out
dither it would be 4 times a month on my unit. That will allow a very large
filter TC.
Bert Kehren
In a message dated 1/30/2012 1:40:57 P.M. Eastern Standard Time,
albertson.chris at gmail.com writes:
On Mon, Jan 30, 2012 at 9:16 AM, <EWKehren at aol.com> wrote:
> With a Rb we are talking loop time in excess of 20 minutes. Any thing
other
> than digital is out of reach. To day I should get my second Rb and I
will
> put it on aging test.
My guess about why you don't see aging or so little of it is that
maybe some engineer at FEI already did the test and added firmware in
the FE5680 to correct the output for estimated aging. So what you are
measuring is the difference between true aging and the unit's internal
aging model.
Trimble does something like this in holdover mode, the controller
compensates for the OCXO's estimated aging
Chris Albertson
Redondo Beach, California