Bob,
Sorry, I'm not sure, if I've understood the issue - what exactly did you
wrong?
Thank you
Volker
Am 08.07.2013 14:48, schrieb Robert Darby:
> John,
>
> After a night's sleep and a rereading of your post I finally realized
> what I was doing wrong. I did not understand the the role of the
> sampling interval setting and the display rate setting on the 5370.
> When I follow the process below the results are totally consistent.
>
> Sorry to taken your time and thanks all for your help. Now to find
> the original issue.....
>
> Bob Darby
>
> On 7/7/2013 8:57 PM, John Miles wrote:
>> So the only difference between the test setups is the setting of the
>> Display Rate control on the 5370, correct? You're allowing TimeLab to
>> estimate the sample rate automatically, and giving it enough time to
>> converge on a stable reading before hitting 'Start Measurement'?
>> You're correct in that changing the real-world sample rate should
>> yield results that are identical (or at least very similar) to
>> resampling the phase data after the fact. In frequency mode, dead
>> time between readings would make that an iffy proposition, but for
>> data taken in TI mode the outcomes should be close. -- john, KE5FX
>> Miles Design LLC
>