[time-nuts] 3GHZ Extender for HP 53131A and 53132A

XPMUser arnold.tibus at gmx.de
Tue Nov 19 15:19:24 EST 2013


Hello Nigel,
and hello Rick,

thank you for the hint to and the remarks concerning
the 3 GHz extender from SQ5ESM.

I ordered one unit and got it within a few days last week.
I am very happy with it in my 53132A, looks like the original
from HP in detail. Also the manufacturing quality seem to
be of high grade, worth the money.

I checked the sensitivity using my 8663A up to 2.5 GHz(and
Trimble Thunderbolt for the reference frequency). The service
manual says that the reading shall be stable up to -27 dBm.
My unit is stable up to -28 dBm to -29 dBm, -30 dBm is too
low. So the unit does work very fine and behaves as decribed
by HP, does seem to be a realy good work using quality parts
as stated by the seller.
I have no idea why yours is showing different values. Did you
check it again?
I hope the original design by HP does meet the criteria
mentioned by Rick ;-)

Thanks again for all the very interesting comments,

regards,

Arnold



Am 09.11.2013 01:03, schrieb Richard Karlquist:
> On 2013-11-08 15:49, GandalfG8 at aol.com wrote:
> 
>>
>> At 1000 MHz, the highest frequency I can generate right now, I've 
>> measured
>> the channel 3 input sensitivity as -50dBm with a sinusoidal  signal.
>>
>> Regards
>>
>> Nigel
>> GM8PZR
>>
> 
> This high sensitivity is probably a bad thing, not a good thing.
> It is indicative of a dynamic divider.  For a frequency counter
> prescaler, you want a static divider, such as the HP5386 used.
> Dynamic dividers make errors if the signal being measured
> has a broadband noise floor or sufficiently high spurs at any
> frequency.
> 
> Rick Karlquist N6RK
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.


More information about the time-nuts mailing list