[time-nuts] DMTD: Mixer DC offset will result in time offset at zero-crossing detector out?
Bruce Griffiths
bruce.griffiths at xtra.co.nz
Fri Nov 22 12:29:01 EST 2013
Stephan Sandenbergh wrote:
> Hi,
>
> Thanks - mystery solved. This is one of the systems that I looked at,
> and missed the DC block in the second amplification stage. I guess it is
> possibly a large Ceramic 10uF. My bad.
>
> Thank you for putting up those web pages I find them to be very good
> references. I spent quite a lot of time reading through them.
>
> Something that puzzles me though is your mixer termination (
> http://www.ko4bb.com/~bruce/LowNoiseMixerPreamp.html). What is the logic in
> having the second balun (and connected in that way)?
>
>
1st section is common mode low pass filter, 2nd section is differential
common mode low pass filter.
Bruce
> Regards,
>
> Stephan.
>
>
> On 22 November 2013 13:15, Bruce Griffiths<bruce.griffiths at xtra.co.nz>wrote:
>
>
>> Stephan Sandenbergh wrote:
>>
>>
>>> Hi,
>>>
>>> I'm playing with dual-mixer time difference stuff again. And, came across
>>> this and I find it somewhat puzzling since no one else seems to have
>>> encountered it. Possibly because I'm missing something?
>>>
>>> The doubly balanced mixers (of the type known to be used in DMTDs and
>>> phase
>>> noise measurement systems) are known to have DC offsets. So much so that
>>> the guys doing phase noise measurements employ elaborate DC removal
>>> circuits in their preamps to combat this.
>>>
>>> Here's my question: why isn't this DC offset removed in any DMTD circuits
>>> I've seen? It seems standard practice to attach the filtered mixer output
>>> directly to the zero crossing detector.
>>>
>>> I did a quick simulation (see attached):
>>>
>>> The mixer beat is a 10Hz sine 0.7Vpp. If you then use a Collins style zero
>>> crossing detector the first stage will have a small gain (I chose a gain
>>> of
>>> 2.83 from Bruce Griffiths pages (
>>> http://www.ko4bb.com/~bruce/ZeroCrossingDetectors.html)). I then compare
>>> this ideal signal to that of a similar one that is offset by 40mV. Notice
>>> the asymmetry in the signal due to offset.
>>>
>>> 40mV result in 1.8ms offset
>>> 4mV result in 180us offset
>>>
>>> Obviously, once the time offset is there no amount of subsequent slope
>>> amplification will remove it.
>>>
>>> I've tested this in practice and bingo, I now have a very accurate way of
>>> plotting relative mixer DC offset over time.
>>>
>>> Any comments?
>>>
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>>> mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>>
>> One can always add AC coupling to eliminate this effect as in
>> http://www.wriley.com/A%20Small%20DMTD%20System.pdf
>>
>> Bruce
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>> mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
More information about the time-nuts
mailing list