[time-nuts] HP 5061A problem
cdelect at juno.com
cdelect at juno.com
Wed Oct 9 13:23:20 EDT 2013
Adrian,
You say it does not lock.
When you put in operate and push the logic reset button does the green
light come on?
What is the 2nd harmonic meter reading?
If too low the logic will not turn on the green light. Raise the reading
via the A7 gain adjust thru the front panel hole.
Is the green light burnt out?
In operate and looking at the control voltage on the meter does the
control voltage "follow" a slow adjustment back and forth of the fine
Quartz adjust?
Corby