[time-nuts] schematics of frequency counter
Bob Camp
kb8tq at n1k.org
Sun Dec 28 15:35:12 EST 2014
Hi
> On Dec 28, 2014, at 9:19 AM, Li Ang <lllaaa at gmail.com> wrote:
>
> Hi Bob,
> I did some test according to your suggestions. DUT is a symmetricom x72
> rb oscillator. Also, I've tried signal generator as the DUT. R&S SMY01 is
> not as good as HP8662A but that the best I've got. The signal geneator is
> also using FE5650 as ref clock.
>
> According to my test with the TDC today, this unit is not producing very
> stable data.
> I don't have accurate pulse generator, so this is how I test the TDC:
> 0) power the board with battery.
> 1) use FPGA to generate time pulse:
> reg [15:0] shift;
> always @(posedge refclk10M) begin
> shift <= {shift[14:0], sw_gate};
> end
> assign tdc_start = shift[3];
> assign tdc_stop1 = shift[5];
>
> 2) use MCU to pull down sw_gate, the FPGA sync it to refclk10M domain and
> generate input signal for TDC.
>
> 3) use TDC to test the time betwen tdc_start and tdc_stop1
>
> The result is in tdc_test.zip. number * 100ns = time between tdc_start and
> tdc_stop1. (TDC highspeed clock is refclk10M/2).
>
> There 2 issues from the test:
> 1) As we can see from the data, the number is around 1.98x not 2.00x. So
> there is about 2ns delay between tdc_start and tdc_stop1 for this simple
> test code. If it is from the PCB trace and something inside FPGA, this part
> should be a constant value at certain temperature.
So far all correct. If you are using Quartus, you can fire up the timing analyzer and take a look at what it guesses for timing / delay on the pulse. It is not a perfect number, but I’d bet it will confirm that the 2 ns does come from the FPGA.
> I can calculate it by
> measuring 2 cycles and 3 cylces. My current code has not implement this
> part, it should provide some improvement. 2ns time error for 1s gate, that
> is something.
The delay probably is from the input / output fabric on the FPGA ( = output driver). The test you propose should demonstrate this.
> 2) For a 90ps TDC, I think the result should be something like +-0.001
> cycle. But I get something like +-0.003 cycle. I do not know the reason for
> now.
Two reasonable *guesses* would be crosstalk and noise.
1) If there are any other clocks running around during this test, I’d see if they can be shut down. Things like an free running OCXO are good for this - they are easy to isolate.
2) Noise could be internal to the TDC. If it’s 90 ns at one sigma, then you will indeed see +’/- 3 X that (or more) depending on how long you watch it. At least by my math the one sigma on the data is 149 ps. That’s a bit over 90 ps, but not terribly far.
Delaying the signals relative to each other (clock and output) as Magnus suggests in another post is probably a real good idea for sorting some of this out.
Bob
>
>
>
>
> 2014-12-27 22:58 GMT+08:00 Bob Camp <kb8tq at n1k.org>:
>
>> Hi
>>
>> (In reply to several posts. It’s easier for me this way)
>>
>> Ok, that’s good news !!! (and useful data)
>>
>> Your counter performance degraded a bit when you put in 5 db and not much
>> when you put in 8 db.
>>
>> It’s also maybe *too* good news. I suspect that cross talk between the
>> channels may be impacting your results.
>>
>> Next step is to try it with two independent sources and a bit more
>> attenuation. When you try it with two sources, you need to attenuate first
>> one source and then switch the attenuators to the other source. That will
>> help you see if crosstalk from one channel is more of a problem than from
>> the other channel.
>>
>> One parts hint:
>>
>> Cable TV attenuators are much cheaper than their fancy 50 ohm MIniCircuits
>> cousins. They are also something you can pick up down at the corner
>> electronics store. For this sort of testing they are perfectly fine to use.
>> At this point in the testing the mismatch between 75 ohms and 50 ohms is
>> not a big deal. You will need to adapt connectors, but you probably still
>> will save money.
>>
>> =======
>>
>> Op-amps that have enough bandwidth and performance for a high input
>> impedance counter input are rare items. They also are not cheap. Often they
>> come as some sort of current feedback part with low(er) input impedance. If
>> you want your counter to work to 300 MHz, it should accept a 300 MHz square
>> wave. That might mean passing the third or even the fifth harmonic of the
>> square wave. An input channel with 900 or 1500 MHz bandwidth is quite a
>> challenge.
>>
>> One very simple solution is to just grab a high speed comparator like the
>> one used by Fluke / Pendulum (ADCMP565). Drive it directly with your input
>> or clock. Make it your front end device. That’s not an ideal solution, but
>> it will give you the bandwidth and a reasonable input impedance. It
>> requires messy things like a negative supply or a “fake” ground (so would
>> the op amp). It also has an ECL output that needs to be converted to match
>> your FPGA ( hint: use the clock inputs, they are LVPECL compatible).
>> Driving into the FPGA with a differential signal is probably needed to
>> reduce crosstalk.
>>
>> No matter how you do it, input channels are *not* an easy thing to do
>> properly. Even on commercial counters, they often are easy to fool.
>> Designing one is only the start. Fully testing it is equally complex.
>>
>> =========
>>
>> Do not underrate your skills in any way. You are doing far more on this
>> project than any of the rest of the list members have done. We have talked
>> and talked forever about these chips. We talk a lot about these ideas. We
>> suggest lots of complex solutions to various possible problems (like the
>> expensive comparator I suggested above). What we almost never do is
>> actually build a counter. If we build something we don’t fully test it. I
>> have never seen any list member share their results the way you have. I
>> suspect that most of us (yes this includes me) are a bit to scared of the
>> response.
>>
>> Please do not stop your work. Keep letting us know how it is going. This
>> is very exciting !!!
>>
>> Bob
>>
>>> On Dec 27, 2014, at 8:22 AM, Li Ang <lllaaa at gmail.com> wrote:
>>>
>>> Hi Bob,
>>> Here is the data and test scheme.
>>> It does not show much difference.
>>>
>>> 2014-12-26 22:12 GMT+08:00 Bob Camp <kb8tq at n1k.org>:
>>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
> <1228.gif><1228.zip><tdc_test.zip>_______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the time-nuts
mailing list