Hi Magnus,
I'm not familar with error analysis and statistics, can you tell me how
to calculate the jitter with my data? Can you tell me some articles or
tutorials about the calculation that a time-nut usually use? I want to
learn stuffs. :)
Thanks.
2014-12-29 21:58 GMT+08:00 Magnus Danielson <magnus at rubidium.dyndns.org>:
> Hi,
>
> Darn, not reading all the notes. Again.
>
> Well, in that case, scaling should be done... then you get average of
> 198,5075 ns and 149,8 ps RMS jitter, with 1,1 ns peak-to-peak.
>
> The jitter is okish then, but a little better would indeed be nice.
>
> Cheers,
> Magnus
>
>
> On 12/29/2014 01:55 PM, Li Ang wrote:
>
>> Hi Magnus,
>> The unit of these data is not ns but reference clock cycles (100ns).
>> TDC_GP22 measures the time between the edge of tdc_start and
>> tdc_stop1, then it measures the reference clock automaticly. The result
>> you
>> get from it is the ratio of them.
>>
>> 2014-12-29 19:58 GMT+08:00 Magnus Danielson <magnus at rubidium.dyndns.org>:
>>
>> Hi,
>>>
>>> Some quick statistic-processing.
>>>
>>> Histogram of your data:
>>> 1.979 0
>>> 1.980 2
>>> 1.981 46
>>> 1.982 173
>>> 1.983 523
>>> 1.984 1031
>>> 1.985 1301
>>> 1.986 1131
>>> 1.987 648
>>> 1.988 236
>>> 1.989 8
>>> 1.990 1
>>> 1.991 0
>>>
>>> The total sample count is 5100 (wc -l only gives 5099 since there is a
>>> missing end of line, but word count is 5100).
>>>
>>> The average is about 1,985075 ns and it is reasonably gaussian, but with
>>> some systematics, notice how the slope is more abrupt on the higher end
>>> than the lower end. A quick and dirty spreadsheet gives me about 2,243 ps
>>> RMS jitter, which isn't all that bad. Yes, it will spread out to that 11
>>> ps
>>> peak-to-peak jitter, but it is to be expected.
>>>
>>> It's pretty respectable for a home-built.
>>>
>>> Cheers,
>>> Magnus
>>>
>>>
>>> On 12/28/2014 03:19 PM, Li Ang wrote:
>>>
>>> Hi Bob,
>>>> I did some test according to your suggestions. DUT is a symmetricom
>>>> x72
>>>> rb oscillator. Also, I've tried signal generator as the DUT. R&S SMY01
>>>> is
>>>> not as good as HP8662A but that the best I've got. The signal geneator
>>>> is
>>>> also using FE5650 as ref clock.
>>>>
>>>> According to my test with the TDC today, this unit is not producing
>>>> very
>>>> stable data.
>>>> I don't have accurate pulse generator, so this is how I test the
>>>> TDC:
>>>> 0) power the board with battery.
>>>> 1) use FPGA to generate time pulse:
>>>> reg [15:0] shift;
>>>> always @(posedge refclk10M) begin
>>>> shift <= {shift[14:0], sw_gate};
>>>> end
>>>> assign tdc_start = shift[3];
>>>> assign tdc_stop1 = shift[5];
>>>>
>>>> 2) use MCU to pull down sw_gate, the FPGA sync it to refclk10M domain
>>>> and
>>>> generate input signal for TDC.
>>>>
>>>> 3) use TDC to test the time betwen tdc_start and tdc_stop1
>>>>
>>>> The result is in tdc_test.zip. number * 100ns = time between tdc_start
>>>> and
>>>> tdc_stop1. (TDC highspeed clock is refclk10M/2).
>>>>
>>>> There 2 issues from the test:
>>>> 1) As we can see from the data, the number is around 1.98x not 2.00x. So
>>>> there is about 2ns delay between tdc_start and tdc_stop1 for this simple
>>>> test code. If it is from the PCB trace and something inside FPGA, this
>>>> part
>>>> should be a constant value at certain temperature. I can calculate it by
>>>> measuring 2 cycles and 3 cylces. My current code has not implement this
>>>> part, it should provide some improvement. 2ns time error for 1s gate,
>>>> that
>>>> is something.
>>>> 2) For a 90ps TDC, I think the result should be something like +-0.001
>>>> cycle. But I get something like +-0.003 cycle. I do not know the reason
>>>> for
>>>> now.
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> 2014-12-27 22:58 GMT+08:00 Bob Camp <kb8tq at n1k.org>:
>>>>
>>>> Hi
>>>>
>>>>>
>>>>> (In reply to several posts. It’s easier for me this way)
>>>>>
>>>>> Ok, that’s good news !!! (and useful data)
>>>>>
>>>>> Your counter performance degraded a bit when you put in 5 db and not
>>>>> much
>>>>> when you put in 8 db.
>>>>>
>>>>> It’s also maybe *too* good news. I suspect that cross talk between the
>>>>> channels may be impacting your results.
>>>>>
>>>>> Next step is to try it with two independent sources and a bit more
>>>>> attenuation. When you try it with two sources, you need to attenuate
>>>>> first
>>>>> one source and then switch the attenuators to the other source. That
>>>>> will
>>>>> help you see if crosstalk from one channel is more of a problem than
>>>>> from
>>>>> the other channel.
>>>>>
>>>>> One parts hint:
>>>>>
>>>>> Cable TV attenuators are much cheaper than their fancy 50 ohm
>>>>> MIniCircuits
>>>>> cousins. They are also something you can pick up down at the corner
>>>>> electronics store. For this sort of testing they are perfectly fine to
>>>>> use.
>>>>> At this point in the testing the mismatch between 75 ohms and 50 ohms
>>>>> is
>>>>> not a big deal. You will need to adapt connectors, but you probably
>>>>> still
>>>>> will save money.
>>>>>
>>>>> =======
>>>>>
>>>>> Op-amps that have enough bandwidth and performance for a high input
>>>>> impedance counter input are rare items. They also are not cheap. Often
>>>>> they
>>>>> come as some sort of current feedback part with low(er) input
>>>>> impedance.
>>>>> If
>>>>> you want your counter to work to 300 MHz, it should accept a 300 MHz
>>>>> square
>>>>> wave. That might mean passing the third or even the fifth harmonic of
>>>>> the
>>>>> square wave. An input channel with 900 or 1500 MHz bandwidth is quite a
>>>>> challenge.
>>>>>
>>>>> One very simple solution is to just grab a high speed comparator like
>>>>> the
>>>>> one used by Fluke / Pendulum (ADCMP565). Drive it directly with your
>>>>> input
>>>>> or clock. Make it your front end device. That’s not an ideal solution,
>>>>> but
>>>>> it will give you the bandwidth and a reasonable input impedance. It
>>>>> requires messy things like a negative supply or a “fake” ground (so
>>>>> would
>>>>> the op amp). It also has an ECL output that needs to be converted to
>>>>> match
>>>>> your FPGA ( hint: use the clock inputs, they are LVPECL compatible).
>>>>> Driving into the FPGA with a differential signal is probably needed to
>>>>> reduce crosstalk.
>>>>>
>>>>> No matter how you do it, input channels are *not* an easy thing to do
>>>>> properly. Even on commercial counters, they often are easy to fool.
>>>>> Designing one is only the start. Fully testing it is equally complex.
>>>>>
>>>>> =========
>>>>>
>>>>> Do not underrate your skills in any way. You are doing far more on this
>>>>> project than any of the rest of the list members have done. We have
>>>>> talked
>>>>> and talked forever about these chips. We talk a lot about these ideas.
>>>>> We
>>>>> suggest lots of complex solutions to various possible problems (like
>>>>> the
>>>>> expensive comparator I suggested above). What we almost never do is
>>>>> actually build a counter. If we build something we don’t fully test
>>>>> it. I
>>>>> have never seen any list member share their results the way you have. I
>>>>> suspect that most of us (yes this includes me) are a bit to scared of
>>>>> the
>>>>> response.
>>>>>
>>>>> Please do not stop your work. Keep letting us know how it is going.
>>>>> This
>>>>> is very exciting !!!
>>>>>
>>>>> Bob
>>>>>
>>>>> On Dec 27, 2014, at 8:22 AM, Li Ang <lllaaa at gmail.com> wrote:
>>>>>
>>>>>>
>>>>>> Hi Bob,
>>>>>> Here is the data and test scheme.
>>>>>> It does not show much difference.
>>>>>>
>>>>>> 2014-12-26 22:12 GMT+08:00 Bob Camp <kb8tq at n1k.org>:
>>>>>>
>>>>>>
>>>>>> _______________________________________________
>>>>> time-nuts mailing list -- time-nuts at febo.com
>>>>> To unsubscribe, go to
>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>> and follow the instructions there.
>>>>>
>>>>>
>>>>>
>>>>> _______________________________________________
>>>>> time-nuts mailing list -- time-nuts at febo.com
>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>>>>> mailman/listinfo/time-nuts
>>>>> and follow the instructions there.
>>>>>
>>>>> _______________________________________________
>>>>
>>> time-nuts mailing list -- time-nuts at febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>>> mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>> mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/
> mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the time-nuts
mailing list