[time-nuts] Changing ADEV, (was Phase, One edge or two?)
Bob Camp
kb8tq at n1k.org
Sat Oct 25 14:11:51 EDT 2014
Hi
If you plot the data versus the six sigma bars (should have been more precise), you will see it walk outside them as well.
Bob
> On Oct 25, 2014, at 1:06 PM, Tom Van Baak <tvb at LeapSecond.com> wrote:
>
>> In the case of the TimePod, the data can be presented when you have *very* few samples to work with.
>> That said, it is interesting to watch it bring up error bars (which are indeed correctly calculated)
>> and then see the trace walk outside those error bars as the run progresses.
>> There are other measurements that are a bit less susceptible to this.
>> None of them have any magic to get around sqrt(N).
>>
>> Bob
>
> Maybe I don't understand error bars. For a dynamic display like TimeLab, walking outside (1 sigma) error bars is expected about 1/3 of the time, no?
>
> /tvb
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the time-nuts
mailing list