[time-nuts] Once again about counter calibration
Scott McGrath
scmcgrath at gmail.com
Sat Apr 25 15:18:21 EDT 2015
Or feed both oscillators into a mixer and extract the difference which will be the offset of your DUT from your standard
Content by Scott
Typos by Siri
> On Apr 24, 2015, at 10:36 PM, Brent Gordon <time-nuts at adobe-labs.com> wrote:
>
> Just feed the PPS into the scope trigger on step 2.
>
> The low repetition rate on the PPS would make this difficult on an analog scope; on a digital scope it is easy. You can use any sub-harmonic that is less than or equal the counter reference frequency (10 MHz).
>
> Brent
>
>> On 4/24/2015 3:45 PM, d0ct0r wrote:
>>
>>
>> Hello,
>>
>> The input: HP 5386A which I would like to calibrate, Well warmed Tremble
>> Thunderbolt (1PPS only), 10 Mhz Datum OCXO (unknown accuracy), Rigol
>> 1101E Oscilloscope.
>>
>> The goal is to calibrate counter to read the Datum OCXO.
>>
>> ===========================
>>
>> Reading the manual for 5386A, there is simple schema for calibration:
>>
>> 1. Connect HP 5386A 10Mhz OUTPUT to Oscilloscope
>> 2. Connect Frequency Standard to Ext. Trigger on Oscilloscope
>> 3. Adjust the frequency on 5386A TCXO for minimum sideways movement of
>> 10 Mhz signal
>>
>> However, my Trimble TB 10Mhz output is currently in use. I have only
>> 1PPS signal available. The HP manual do not mention what exactly
>> frequency needs to be on "Frequency Standard" connected to Ext. Trigger.
>> Is there any method/option I could apply to calibrate HP counter ?
>>
>> Thanks !