Hi

A very normal “spiral of death” on the 53132 is the fan going out and the power supply following it a month or two later. The supply apparently does need the fan cooling to keep it happy. Since the fan runs whenever the counter is plugged in, the fan logs a lot of hours. If you do loose the power supply, the one in the 53181 will swap over to the 132. It’s not a cheap fix, but better than scrapping a 132.

Bob

> On Jan 11, 2015, at 12:59 AM, Loïc Moreau <loic.moreau at eai.fr> wrote:
> 
> Hi,
> My 53132A cooling fan is becoming extremely noisy so I have to fix it or find ear plugs fast.
> 
> If anybody as encountered the same problem witch I suppose is common I will be interested if there is replacement part easily available for that purpose.
> 
> Regards
> Loïc
> 