My understanding is that MVAR(m*tau0) is equivalent to filtering the phase
samples x(n) by averaging m samples to produce x'(n)
[x'(n)=1/m*(x(n)+x(n+1)..x(n+m-1))] and then calculating AVAR for
tau=m*tau0 on the filtered sequence. Thus, MVAR already performs an
averaging/ lowpass filtering operation. Adding another averaging filter
prior to calculating MVAR would seem to be defining a new type of stability
measurement.

Not familiar with the 5370... Is it possible to configure it to average
measurements over the complete tau0 interval with no dead time between
measurements? Assuming the 5370 can average 100 evenly spaced measurements
within the measurement interval (1s?), calculating MVAR on the captured
sequence would produce MVAR(m*.01)) for m being a multiple of 100. i.e.,
tau0 here is actually .01, not 1, but values for MVAR(tau) for tau's less
than 1s are not available.

Shouldn't the quantization/ measurement noise power be easy to measure?
Can't it just be subtracted from the MVAR plot? I've done this with AVAR in
the past to produce 'seemingly' meaningful results (i.e. I'm not an expert).

I calculated the PSD of x(n) and it was clear where the measurements were
being limited by noise (flat section at higher frequencies). From this I
was able to estimate the measurement noise power.

AVAR_MEASURED(tau)=AVAR_CUT(tau)+AVAR_REF(tau)+AVAR_MEAS(tau)

i.e. The measured AVAR is equal to the sum of the AVAR of the clock under
test (CUT), the AVAR of the reference clock, and the AVAR of the
measurement noise. If the reference clock is much better than the CUT
AVAR_REF(tau) can be ignored. AVAR_MEAS(tau) is known from the PSD of x(n)
and can be subtracted from AVAR_MEASURED(tau) to produce a better estimate
of AVAR_CUT(tau).

Depending on the confidence intervals of AVAR_MEASURED(tau) and the noise
power estimate, you can get varying degrees of cancellation. 10dB of
improvement seemed quite easy to obtain.

James


Message: 7
> Date: Tue, 28 Jul 2015 21:51:07 +0000
> From: Poul-Henning Kamp <phk at phk.freebsd.dk>
> To: time-nuts at febo.com
> Subject: [time-nuts] Modified Allan Deviation and counter averaging
> Message-ID: <2884.1438120267 at critter.freebsd.dk>
> Content-Type: text/plain; charset="us-ascii"
>
> Sorry this is a bit long-ish, but I figure I'm saving time putting
> in all the details up front.
>
> The canonical time-nut way to set up a MVAR measurement is to feed
> two sources to a HP5370 and measure the time interval between their
> zero crossings often enough to resolve any phase ambiguities caused
> by frequency differences.
>
> The computer unfolds the phase wrap-arounds, and calculates the
> MVAR using the measurement rate, typically 100, 10 or 1 Hz, as the
> minimum Tau.
>
> However, the HP5370 has noise-floor in the low picoseconds, which
> creates the well known diagonal left bound on what we can measure
> this way.
>
> So it is tempting to do this instead:
>
> Every measurement period, we let the HP5370 do a burst of 100
> measurements[*] and feed the average to MVAR, and push the diagonal
> line an order of magnitude (sqrt(100)) further down.
>
> At its specified rate, the HP5370 will take 1/30th of a second to
> do a 100 sample average measurement.
>
> If we are measuring once each second, that's only 3% of the Tau.
>
> No measurement is ever instantaneous, simply because the two zero
> crossings are not happening right at the mesurement epoch.
>
> If I measure two 10MHz signals the canonical way, the first zero
> crossing could come as late as 100(+epsilon) nanoseconds after the
> epoch, and the second as much as 100(+epsilon) nanoseconds later.
>
> An actual point of the measurement doesn't even exist, but picking
> with the midpoint we get an average delay of 75ns, worst case 150ns.
>
> That works out to one part in 13 million which is a lot less than 3%,
> but certainly not zero as the MVAR formula pressume.
>
> Eyeballing it, 3% is well below the reproducibility I see on MVAR
> measurements, and I have therefore waved the method and result
> through, without a formal proof.
>
> However, I have very carefully made sure to never show anybody
> any of these plots because of the lack of proof.
>
> Thanks to Johns Turbo-5370 we can do burst measurements at much
> higher rates than 3000/s, and thus potentially push the diagonal
> limit more than a decade to the left, while still doing minimum
> violence to the mathematical assumptions under MVAR.
>
> [*] The footnote is this: The HP5370 firwmare does not make triggered
> bust averages an easy measurement, but we can change that, in
> particular with Johns Turbo-5370.
>
> But before I attempt to do that, I would appreciate if a couple of
> the more math-savy time-nuts could ponder the soundness of the
> concept.
>
> Apart from the delayed measurement point, I have not been able
> to identify any issues.
>
> The frequency spectrum filtered out by the averaging is waaaay to
> the left of our minimum Tau.
>
> Phase wrap-around inside bursts can be detected and unfolded
> in the processing.
>
> Am I overlooking anything ?
>
>
> --