[time-nuts] ADEV noise floor vs counter gate time

Dave Martindale dave.martindale at gmail.com
Wed Mar 18 11:22:03 EDT 2015


The counter always has a 1 count uncertainty in the timebase measurement,
which is a 2e-8 error with a 1 second gate time.  If the value being
displayed starts with the digit 9, and 8 digits are displayed, then that
translates to +- 2 counts in the last place.  But the measurements are
synchronized to the input signal, so it always measures an integer number
of input cycles, and there should be no comparable uncertainty in the input
(other than some noise in deciding exactly when the edge crosses the input
threshold, which should be tiny compared to the 20 ns timebase period).

But that's comparing the counter reading to the real world.  My table was
comparing the displayed output to the likely raw measurements, and it seems
to show that the counter's internal math is being performed to the full 10
digits of precision in the USB data, even when the gate time supports only
8 bits of accuracy.  That's good news because it allows you to know when
you have correctly guessed the input counts.

When trying to calculate the raw data from the reading, you do need to try
an input count of 91 in addition to 90 and 92.  91 did show up in the small
sample of period-mode measurements, even if it did not appear in any of the
frequency-mode measurements.

I don't think the counter is doing "averaging", in the sense of making
multiple independent short-period measurements and then averaging them for
higher precision.  Instead, it is just making one long continuous
measurement, sampling the signal periodically, and then actually
calculating frequency or period from two measurements separated by an
appropriate time.  For a simplified example:

Suppose the counter generates a time stamp approximately every 1 second
(always aligned with the input signal active edge) and then stores the two
pieces of raw data (the current input cycle counter and the current
timebase counter) in a small memory buffer.  The counters are never reset;
they just need to be large enough to never overflow twice within the
longest input period allowed (1000 s for the TF930).  To display a
frequency or period based on a 1 s gate time, the counter simply subtracts
two successive data records to get delta-input and delta-timebase counts,
then does its calculations based on those deltas.  To display a 10 second
gate time measurement, the counter looks back through its memory to find a
time stamp about 10 seconds earlier than the most recent measurement (with
high input frequency, that will generally be 10 measurements ago, but when
measuring a signal with a 0.2 Hz frequency it's only 2 measurements ago).
For a 100 second gate time measurement, the counter needs to find a saved
time stamp that is about 100 seconds ago.  Once it has found the correct
data record, it calculates the difference in input and timebase counts
between that one previous data record and the most recent, and then
calculates the displayed value from it.

One second later, the counter can calculate a new 100 s measurement, using
the new data point just captured and a different stored data point 100
seconds ago.  But 99 of the 100 seconds in the new gate period are shared
with the old gate period, so the displayed value is not likely to change
very much simply because 99% of the observation period is shared.

Thus, every displayed value is calculated from only 2 time-stamped
measurements.  The longer gate time places those measurements further
apart, reducing the uncertainty due to the +- 1 clock of the timebase.
Because the counters run continuously without resetting, no clock edges are
lost and a 100 s gate time measurement has only 20 ns uncertainty in the
whole 100 s period.  Also, any wander in the input frequency between those
two measurements is invisible if it cancels out over the gate time.  So
there is "averaging" in the sense that the counter display always reflects
what is happening on the scale of the gate time, but it's not computing and
then averaging multiple numbers.

I have not tried doing my own ADEV calculations, so I can't say what it is
about the shorter gate periods that make the oscillator appear noisier than
it really is.  How does the ADEV calculation aggregate a stream of
short-time calculations into measurements for large tau values?  My
intuition is this:  If you just take readings from the counter with a 1 s
gate time, each reading has a 2e-8 uncertainty.  If you average a bunch of
these readings, and the errors are independent, the accuracy should improve
by a factor of sqrt(N).  But if you unwrap each reading into an integer
number of input and timebase cycles, you essentially have a series of phase
samples that can be added or subtracted without increasing the absolute
uncertainty.  So when you combine 100 1 second measurements, you get a
relative accuracy that is 100 times better, not the sqrt(100) you'd get
from averaging.

- Dave


On Wed, Mar 18, 2015 at 6:33 AM, <jpbridge at aol.com> wrote:

> Hi Dave,
>
> Interesting analysis. The accuracy stated in the manual is ...+ 2 counts
> and though this relates to the 50MHz clock, perhaps they use a similar
> algorithm for the input frequency.
>
> I completed the 0.3 second measurements and the curve is similar to 1
> second but higher up (i.e. as you'd expect by extrapolation from the
> behaviour of the other curves).
>
> My ADEV calculation is based on the average frequency in each bin, the
> varying size of the bin should be insignificant as long as it is not
> affecting the average value within the bin. If the average frequency shifts
> by delta_F in one bin time step and the first bin is delta_T short (as a
> fraction of one bin time step) then the first frequency will be
> delta_T*delta_F low and the second bin perhaps that much high but the key
> point is that it is the product of the two deltas so it won't materially
> affect the accuracy of the calculation. At least I think that is correct.
>
> Taking the worst possible case where the delta in bin size always went the
> wrong way so every term in the Alan Variance sum was multiplied by
> (1+2delta)^2 then the final Alan deviation might be (1 + 2 delta) too big
> but as delta is of the order of 10E-8 or less this wouldn't even register
> on the graphs.
>
> What I might try doing is programming your approach into the code to try
> and get at the raw data - I only need to try 88,90 and 92 as possible
> counts - though to be sure I'll try mean frequency +- 5 say and then try
> and get the 50MHz clock values out as integers. What I might also do is
> then do a least squares fit (linear regression) to get the frequency over
> each bin and use the slope (this perhaps is what the counter does
> internally - I don't know).
>
> I'd like to get to the bottom of this if only to understand my counter
> better.
>
> James
>
>
>
>
>
>
>
>  -----Original Message-----
> From: Dave Martindale <dave.martindale at gmail.com>
> To: jpbridge <jpbridge at aol.com>; time-nuts <time-nuts at febo.com>
> Sent: Wed, 18 Mar 2015 1:26
> Subject: Re: [time-nuts] ADEV noise floor vs counter gate time
>
>  I believe I see the pattern.  As you figured out, you wouldn't expect a
> single period to be a multiple of 20 ns; you expect the length of (about)
> 90 periods to be an integer multiple of 50 ns, since that's what the
> counter actually measures.  Further, the measuring time isn't exactly 1
> second, it is an integer number of periods of the input frequency that
> makes up at least 1 second.  If the counting logic was all hardware, you
> would expect to capture either 90 or 91 cycles of the input, depending on
> whether the input frequency was slightly below or above 90 Hz respectively.
>
> I built this table of your frequency data in Excel.  Math is 64-bit
> floating point, equivalent to about 16 decimal digits, so plenty accurate
> enough to simulate this counter:
>
> Reading    Input Count TB Count      Rounded      Frequency       Interval
>  90.00006359    92    51111074.998    51111075    90.00006359
> 1.022221500
>  90.00007591    92    51111068.002    51111068    90.00007591
> 1.022221360
>  89.99999640    90    50000002.000    50000002    89.99999640
> 1.000000040
>  89.99998740    90    50000007.000    50000007    89.99998740
> 1.000000140
>  90.00006007    92    51111076.997    51111077    90.00006007
> 1.022221540
>  89.99996040    90    50000022.000    50000022    89.99996040
> 1.000000440
>  90.00008648    92    51111061.999    51111062    90.00008648
> 1.022221240
>  90.00008472    92    51111062.999    51111063    90.00008472
> 1.022221260
>  90.00011465    92    51111046.001    51111046    90.00011465
> 1.022220920
>  90.00014459    92    51111028.998    51111029    90.00014459
> 1.022220580
>
> The first column is your data.  The second column is a guess about how
> many input cycles were captured.  The third column is the number of
> timebase cycles that have elapsed since the previous reading, based on the
> first two columns.  I hand-tweaked the numbers in the second column until
> the number in the third column was within 0.003 of an integer.  The fact
> that I was always able to do this tells me that my guess is probably
> correct, and the small residual (which is a few parts in 1e-10) is due to
> the counter rounding the results to 10 digits.  The 4th column is the
> result of rounding the previous column to the nearest integer.  This is
> what I believe is the actual number of counts the counter saw.  The 5th
> column is a fresh calculation of frequency, based on the integer number of
> input cycles in column 2 and the integer number of timebase cycles in
> column 4.  When the result is rounded to 10 digits, you can see it matches
> the 10 digits that the counter provided back in column 1.
>
> Oddly, the counter never captured 91 input cycles.  If the input frequency
> was a little higher than 90 Hz, it always measured at 92 cycles, even
> though 91 cycles was well more than 1 s since the previous reading.  I
> guess the microprocessor running the counter only checks periodically (e.g.
> every 20 ms) to see if the gate time has elapsed, and then latches the
> counts on the next active edge of the input signal.
>
> So, I claim that with this small sample, at least, we recovered the exact
> number of 20 ns periods between samples, and the number of integer input
> cycles as well.  Also notice the 6th column.  This is the actual sample
> interval, based on the number of elapsed timebase counts.  Note that the
> sample period is *not* exactly 1 second, nor is it even close to a constant
> value, since some measurements are of 90 cycles while others are of 92
> cycles.  Does your ADEV calculation algorithm take into account the
> variable spacing of the input samples in time?  If it assumes they are
> regularly spaced (i.e. every 90 cycles) it may get confused by this
> variable-spacing data.
>
> Now here is almost the same process applied to your period data:
>
> Reading     Input Count  TB Count      Rounded         Period
> Interval
>  0.01111107736    91    50555401.988    50555402    0.01111107736
> 1.011108040
>  0.01111110130    92    51111065.980    51111066    0.01111110130
> 1.022221320
>  0.01111110769    91    50555539.990    50555540    0.01111110769
> 1.011110800
>  0.01111110435    92    51111080.010    51111080    0.01111110435
> 1.022221600
>  0.01111110593    91    50555531.982    50555532    0.01111110593
> 1.011110640
>  0.01111110022    90    49999950.990    49999951    0.01111110022
> 0.999999020
>  0.01111114000    90    50000130.000    50000130    0.01111114000
> 1.000002600
>  0.01111110000    90    49999950.000    49999950    0.01111110000
> 0.999999000
>  0.01111110370    92    51111077.020    51111077    0.01111110370
> 1.022221540
>
> Again, column 2 was hand-adjusted for each row to keep the third column
> close to an integer.  The residual errors here are larger, since the
> maximum rounding error of 0.5 in the last place is a larger change relative
> to a 10-digit value of 11111111 than it is to a value of 90000000, but all
> are still within 0.02 of being an integer.  This time, the counter grabbed
> measurements after 90, 91, or 92 cycles.  Again, after rounding the
> timebase count to an integer and calculating a 10-digit period for display,
> the result always matched what the counter output.  Again, I think we know
> with high probability just how many input and timebase cycles were counted
> for each measurement.
>
> I adjusted column 2 by eye, while looking at the results of column 3, but
> that process could be automated pretty easily (just not in Excel).  As I
> tried 90, 91, and 92 in sequence, there was always just one of those which
> gave a small residual error.
>
> So I think your TF930 is making measurements and accurately converting
> them to frequency or period, with a +- 20 ns uncertainty for each
> measurement.  Since it is a time-stamping counter, the uncertainty in a 10
> s or 100 s or 1000 s measurement time (assembled by external software) is
> still only 20 ns.  That's great, but to actually get that accuracy over a
> long measurement time, you will need to determine and add up the actual
> number of input counts and timebase counts.  And you will have to
> understand that the counter does not make measurements at constant or
> near-constant intervals (e.g. every 90 cycles of input, without
> exception).  It gives you measurements whenever it gets around to measuring
> them.
>
> Too bad there doesn't seem to be a way to get it to return the raw
> observed data (input cycle count, timebase cycle count) instead of the
> frequency or period derived from them.  That would make it trivial to
> string together a bunch of 1s measurements into arbitrarily long gate
> times.
>
> - Dave
>
>  On 17/03/2015 05:57, jpbridge at aol.com wrote:
>
> Hi Dave,
>
> Thank you for your detailed response.
>
> I use the E? command because it returns results at the gate time intervals
> rather than at the LCD update rate (as you point out). I think that this is
> working correctly because I get very different file sizes.
>
> The numbers are returned as strings of 10 digits - here are some for 1
> second gate:
>
>  90.00006359e+0Hz
> 90.00007591e+0Hz
> 89.99999640e+0Hz
> 89.99998740e+0Hz
> 90.00006007e+0Hz
> 89.99996040e+0Hz
> 90.00008648e+0Hz
> 90.00008472e+0Hz
> 90.00011465e+0Hz
> 90.00014459e+0Hz
>
> I generally use the frequency mode but I also tried time period and found
> I got the same curve in essence, which was comforting in a way but showed
> it wasn't rounding in converting to frequency.
>
> The numbers above, on my calculator at least don't exactly match counts of
> 20 nanosecs.
>
> Here are some time period results:
>
> 11.11107736e-3s
> 11.11110130e-3s
> 11.11110769e-3s
> 11.11110435e-3s
> 11.11110593e-3s
> 11.11110022e-3s
> 11.11114000e-3s
> 11.11110000e-3s
> 11.11110370e-3s
>
> Again they don't seem to be integer values of 20 nanosec exactly, though
> quite close. For example
> 11.11107736E-3/20E-9 = 555,553.868
> 555,554 x 20E-9 = 11.11108E-3
>
> But I guess what it returns is the ratio of counts within the gate. So
> 11.11107736E-3 period will occur
> 90 times in a second (as it is slightly short) and so I should take the
> ratio:
>
> 90 x 11.11107736E-3/20e-9 = 49,999,848.12
>
> so still not quite an integer but if I assume the count (of 50MHz periods)
> was 49,999,848 and calculate one 90 th of it I get:
>
> 49,999,848 x 20E-9/90 = 1.1111077333333
>
> Still not exact agreement. I note that .12 is very close to .125 or 1/8
> but I don't know if that is significant.
> It is probable that it rounds the ratio in binary and then converts to
> decimal to print out.
>
> I've tried assuming 89 periods and 91 periods but still don't get exact
> integer ratios.
>
> Anyway, as I get good agreement between period and frequency measurements
> at 1 sec, I don't think that it is a rounding issue.
>
> I do think it is a quantization issue down to the +/- 20 nanosecs/gate
> time but I can't quite work it out.
>
> I'm currently doing a run at 0.3 secs gate time and I'll see what sort of
> curve that produces.
>
> Tomorrow I should receive my new Tek counter (I went for the fca3100 in
> the end as I got a very good discount on an ex demo unit) and that should
> give something to compare (once I've worked out how to program it).
>
> James
>
>
>  -----Original Message-----
> From: Dave Martindale <dave.martindale at gmail.com>
> To: jpbridge <jpbridge at aol.com>; Discussion of precise time and frequency
> measurement <time-nuts at febo.com>
> Sent: Tue, 17 Mar 2015 0:27
> Subject: Re: [time-nuts] ADEV noise floor vs counter gate time
>
>   How is the counter configured?  Are you reading period or frequency?
> Are you in "E?" (Every Result) mode, or "C?" (Continuous Result) mode?  The
> former should give you continuous but independent measurements, while the
> latter gives heavily overlapped measurements.  (For example, with a 100
> second gate time, you get one E output every 100 seconds, which covers a
> different 100-second period than the previous measurement.  In C mode, you
> get one output every 2 seconds, each of which is an estimate from 100
> seconds of measurement, but 98 seconds of that data was also part of the
> previous output and only 2 seconds of new data is included).
>
>  What does the data returned by the counter actually look like?  The
> manual implies that you always get 10 digits worth of result (not including
> the exponent) regardless of gate time, but are the values rounded for
> display in 7, 8, or 9 digits at the shorter gate times, or are they a full
> 10 digits always?  Given any particular value of frequency or period you
> get, you should be able to reverse-calculate the number of whole cycles of
> the input signal that the counter used as a gate time, and the number of
> cycles of 50 MHz timebase that were counted in that period.  Since the
> counter doesn't have interpolators, both of these values should be
> integers, and so the possible output values are a small subset of all
> possible 10-digit values for the shorter gate times.
>
>  For example, if the difference frequency is exactly 90 Hz, the period
> between two "1 second" measurements will be exactly 1 second, and the
> counter will record 90 cycles of input and 5e7 cycles of timebase,
> exactly.  In frequency mode, the output should be 90.0 Hz exactly, and in
> period mode the output should be 11.11111111 ms.  Now suppose that the
> difference frequency is just a hair slow, enough that 90 cycles of input
> spans 50,000,001 counts of the timebase.  The reported frequency should be
> 89.99999820 Hz and the reported period should be 11.11111133 ms.  With a 1
> s gate time, no values between those are possible unless the values are
> being rounded (or there is an error in the calculation, which is always
> possible).  Looked at another way, the smallest possible change in the
> reported period is one timebase clock (20 ns) divided by the number of
> input cycles in one gate time (90 for 1 s).
>
>  If the counter is rounding, you may be able to unambiguously figure out
> what the actual inputs (cycles of input and cycles of timebase) to the
> calculation were, and use that instead of the rounded value in your
> calculations.  Rounding may round up or down, but if the two oscillators
> are stable enough the direction can be predominantly "up" or "down" for
> long periods of time, adding a bias to the actual frequency or period
> you're measuring.  (I don't know what effect this bias would have on ADEV).
>
>  - Dave
>
>  On Mon, Mar 16, 2015 at 10:15 AM, James via time-nuts <time-nuts at febo.com
> > wrote:
>
>> Hi All,
>>
>> I'm in the process of getting a better counter, but at present I'm using
>> my TTi TF930 counter.
>>
>> For those who don't know it, it is a reciprocal counter which should be
>> continuous, it counts periods in terms of its internal 50MHz clock which
>> I've locked to an external 10MHz reference.
>>
>> There are 4 gate times available, 0.3 secs, 1 sec, 10 secs and 100 secs.
>>
>> These correspond to 7, 8, 9 and 10 digits.
>>
>> I've been experimenting with using a single mixer (mini circuits ZAD+)
>> along with a 1MHz low pass filter and appropriate attenuators to measure
>> Alan Deviation (using my own software).
>>
>> My set up is a 10MHz reference source (MV89A which I've approximately set
>> using a 10kHz GPS signal).
>>
>> The reference is used as the external reference for an Agilent 33522A
>> arbitrary waveform generator.
>>
>> The 33522A generates an 9.999910 MHz (10MHz - 90Hz) sine wave at 300mVpp
>> to the mixer and the mixer is also fed by the 10MHz reference output of the
>> 33522A via an attenuator to get it to roughly the same level.
>>
>> The second output of the 33522A generates a 10MHz square wave as a
>> reference for the counter (the counter requires quite a high reference
>> signal and the reference out of the 33522A is too low a voltage to be used
>> directly).
>>
>> I initially ran this with a gate of 1 second and the LOG10(ADEV) curve
>> drops linearly vs LOG10(tau) but then curves back up again. (I tried many
>> variants such as using period rather than frequency and so on.)
>>
>> But when I set the gate time to 10 seconds or 100 seconds then I get both
>> lower curves and ones that no longer curve upwards.
>>
>> The attached pdf shows the three curves on the same graph.
>>
>> What puzzles me is that the counter at longer gates is only averaging to
>> get more digits so the difference must come down to quantization in terms
>> of the number of digits that are passed to the computer over the USB/RS232
>> link.
>>
>> I find it rather puzzling.
>>
>> James
>>
>>
>>
>>
>>
>>