Hi

*IF* I’m understanding you correctly, you are measuring this data with an HP 5386.

That’s only a 4 ns counter. I agree with Tom that there is something weird in your setup, but that
still does not help the basic limit of the counter’s resolution. You will need to be out around 1,000 
seconds or more to get to the noise on the OCXO. At any time shorter than that you are either
measuring the counter’s noise or noise in your setup.

If you have one, a 5335 would be a better counter to use than the 5386. 

Bob

> On May 27, 2015, at 10:04 PM, Vlad <time at patoka.org> wrote:
> 
> 
> Hello,
> 
> Recently, I acquired video distribution amplifier (model Altinex DA1804NT). It says it could handle 350Mhz. I am planing to use it to split 10Mhz Ref signal from my OCXO. This OCXO I bought on auction site. It has Oscilloquarz model 8663-SX plus some components which give me 4 outputs - 2 sines and two squares. After few days of "warm up" I did measurement of ADV from one of the sine outputs and then measurement of the same sine signal from OCXO, but this time through the Altinex DA. Each test took 8 hours. Attached is the results.
> 
> NOTE: on the ALTINEX* charts I forgot to change the label for OCXO. So, its not DATUM. It is Oscilloquarz 8663-SX
> I was using Trimble TB as a ref signal  for HP 5386
> 
> Since I have no  instruments which measuring phase noise, I am interesting, what will be the quick method to measure the phase noise for that DA ?
> 
> 
> -- 
> WBW,
> 
> V.P.<ALTINEX-DA.png><ALTINEX-DA-F.png><ALTINEX-DA-T.png><OCXO2.png>_______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.



More information about the time-nuts mailing list