There are a number of SRS 620’s with big block letter warning sheets taped to the top
of them about error 34. It is a *very* common thing to run into and (obviously) something
the guys in the repair shop get tired of really fast.
Bob
> On Dec 22, 2016, at 1:27 PM, Magnus Danielson <magnus at rubidium.dyndns.org> wrote:
>
> Rick,
>
> On 12/22/2016 05:53 PM, W2GPS wrote:
>> Time-nuts,
>>
>> My SR620 has started reporting "test error 34" on power-up and will do
>> nothing after that. Does anyone know what that message means? I have a
>> reasonable lab and I would try to repair the unit myself but I have no
>> documentation. Is service documentation available or should I just ship it
>> back to Stanford Research to get fixed?
>
> The manual, which you can download for free from SRS, says (page 57):
>
> "If the UHF prescaler of channel A is selected when
> the unit is powered off and then on again, the unit
> will display “TEST ERROR 34”. This error may be
> corrected by deselecting the UHF of channel A. To
> run this procedure, press input button of channel
> “A” until UHF is not highlighted. Then turn the
> power off and on again."
>
> So RTFM and execute accordingly might be all that you need to do.
>