[time-nuts] HP5065A environmental sensitivities
Magnus Danielson
magnus at rubidium.dyndns.org
Sun May 22 12:38:35 EDT 2016
Hi Bob,
On 05/22/2016 05:21 PM, Bob Camp wrote:
> Hi
>
>> On May 22, 2016, at 10:49 AM, Attila Kinali <attila at kinali.ch> wrote:
>>
>> Hoi Poul-Henning,
>>
>> On Fri, 20 May 2016 09:52:39 +0000
>> "Poul-Henning Kamp" <phk at phk.freebsd.dk> wrote:
>>
>>>> Temperature effects are by far the largest effects we have to deal with.
>>>
>>> Which is why you'd be much better of with 20 sensors at 16 bits,
>>> than a single sensor at 24 bits.
>>
>> Does the HP5065 have such huge temperature gradients/differentials inside?
>
> At least 5C offset relative to an external sensor. Likely more in some rack installations. Past
> that it depends on how close the item of interest is to the oven assembly.
To explain a little more.
The 5065A/B has a design that relies on self-convection. Depending on
which generation, the AC rectifier is sitting on the PCB or on the
chassi for cooling. Similarly the switcher on the A15 board can run a
bit hot. This alone makes me recommend that the 5065 is put such that
there is self-convection. I run mine free-standing.
Making the 5065 more temperature stable is definitely recommended, but
that is to some degree orthogonal from handling the heat produced.
PHKs replacement of PSU details helps to reduce the heat and also
produces more stable power-supply, which is a good step in the right
direction on both accounts.
With a little cleanup of the PSU, thermal management have improved
somewhat in the box.
The negative voltage switcher of A15 board should be replaced with a
suitable switcher, and some of the DC/DC switcher modules is built with
isolation, so it should allow for mounting two in parallel such that you
get both sides. PHK, have you checked this option?
Cheers,
Magnus
More information about the time-nuts
mailing list