[time-nuts] uncertainty calculations
Bill Byrom
time at radio.sent.com
Sat Apr 15 21:35:30 EDT 2017
I believe that the problem is that the error in any one measurement is
not uniformly distributed in exactly that way. If you are trying to
count how many 10 MHz (100 ns interval) intervals occur between the 1
PPS edges in a period counter, you have to deal with the following?
* You might have any possible phase relationship between the two
signals. If they are exactly related by a 10^7 ratio, it's possible
for the 1 PPS edges to exactly coincide with the 10 MHz edges.
Depending on the type of gating circuit, you will have jitter and
possibly metastability resolving whether which edge occured first. The
same thing happens on the end of the measured interval, but (depending
on how it's set up) the propagation delays and metastability and
jitter might be different. So you could get millions of sequential
counts which were 1 count low, followed by millions of counts which
were one count high, with no counts exactly at 10^7.
* To stay away from such problems, most precision counters add a small
amount of controlled jitter (phase modulation) to the clock. When
averaged over many measurements the effects of the two edges (gate and
clock) lining up exactly are greatly reduced, since you are sliding
one back and forth across the other with the modulation and the chance
of metastability is small (assuming the signal being measured doesn't
happen to match the phase modulation frequency).
* The metastability problem depends on how the edges are compared. Some
traditional flip-flops and latches can be thought of as analog gain
elements connected so that they tend to sit in state A or state B,
which involve analog voltages and currents. If you graph the energy in
the system, the energy is low in state A, rises to a peak halfway
between A and B, and falls to a low value at state B. If the
recognition of the timing edge occurs early enough the system remains
in state A. If the timing is later the system is pushed toward the
peak, but doesn't get over it and returns to state A. But if the
timing is at the perfect location the system is balanced at the
potential energy peak, and only random noise can push the system into
a final state A or B over a significant length of time.
Sorry if this is considered obvious or trivial.
--
Bill Byrom N5BB
----- Original message -----
From: jimlux <jimlux at earthlink.net>
To: Discussion of precise time and frequency measurement <time-nuts at febo.com>
Subject: Re: [time-nuts] uncertainty calculations
Date: Fri, 14 Apr 2017 08:49:07 -0700
On 4/14/17 8:37 AM, jimlux wrote:
> If one is counting an unknown 1pps source with a counter that
> runs at 10
> MHz (e.g. the error in any one measurement is uniformly
> distributed over
> 1 ppm) and you collect 100 samples,
> is the (1 sigma) measurement uncertainty 0.1ppm * sqrt(100)/sqrt(12)
>
> (standard deviation of a uniform distribution is 1/sqrt(12) )
>
> (assuming for the moment that both sources have no underlying
> variability - we're talking about the *measurement uncertainty*)
>
Oops..
0.1 ppm * 1/sqrt(N) * 1/sqrt(12)
That is, the standard deviation goes down as sqrt(N)