[time-nuts] HP5061B Versus HP5071 Cesium Line Frequencies
Bob kb8tq
kb8tq at n1k.org
Fri Jun 2 20:12:05 EDT 2017
Hi
> On Jun 2, 2017, at 7:45 PM, Donald E. Pauly <trojancowboy at gmail.com> wrote:
>
> https://www.febo.com/pipermail/time-nuts/2017-May/105566.html <https://www.febo.com/pipermail/time-nuts/2017-May/105566.html>
Ok, so yet again a reference to the start of this thread … why?
>
> Electronic thermal coolers did not exist then so it could not be done.
> Electronic temperature sensors did not exist either.
I guess they must have just dreamed up the pelter devices they used. FYI, they have
been around since 1834 (no that’s not a typo).
> That crystal cut
> has been known since the 1940's at least.
And once you get away from an AT or SC, how much is known about the mode spectra of
the cut ….
> It has been neglected
> because of limited temperature range. It yields ±1 ppm over a range of
> ±20° C from 25° C. A slightly different angle of cut can yield ±250
> ppb over that range. (4:1 improvement) Contrast that with a normal AT
> cut which yields ±9 ppm over that range.
Umm …. errr … it’s quite easy to get a +/- 2 ppm 0-50C AT cut *including* the tolerance
on the cut angle.
>
> I built an oven with an Analog Devices temperature sensor 20 years
> ago. I did not have time to incorporate foam insulation. The heater
> power was not available to run it at 65° C without insulation. It had
> to run at 40° C and it would hold about 1 ppb over a few hours. It
> would hold the crystal within 0.01° or so but it was far away from the
> turnover temperature. Convection currents cause problems. It
> convinced me that ovens were headaches. Thermal coolers remove most
> of these.
I’d suggest you try a few more experiments with real crystals in real applications.
Bob
>
> πθ°μΩω±√·Γ
> WB0KVV
>
> ---------- Forwarded message ----------
> From: Bob kb8tq <kb8tq at n1k.org>
> Date: Fri, Jun 2, 2017 at 3:50 PM
> Subject: Re: [time-nuts] HP5061B Versus HP5071 Cesium Line Frequencies
> To: Discussion of precise time and frequency measurement <time-nuts at febo.com>
> Cc: "rward0 at aol.com" <rward0 at aol.com>, "Donald E. Pauly"
> <trojancowboy at gmail.com>
>
>
> Hi
>
> Have you checked out the papers from the 1950 and `1960’s where they
> actually tried what you
> propose with essentially the same parts you are looking at using?
>
> Bob
>
>
>> On Jun 2, 2017, at 5:51 PM, Donald E. Pauly <trojancowboy at gmail.com> wrote:
>>
>> # 2 is not true. A cut has either two turning points or zero. Where
>> both turning points exist there are two temperatures at which the
>> temperature coefficient of frequency is zero. Cut 0 on figure 6 at
>> https://coloradocrystal.com/applications has no turnover point. It is
>> neither fish nor fowl. Cut 6 is the normal AT curve with extremes of
>> ±16 ppm for -55° C thru +105° C. All curves normally intersect at 25°
>> C rather than the 27° C shown. 25° C is half way between -55° C thru
>> +105° C. Curve 6 is the Tchebychev polynomial y=4x^3-3x and curve 0
>> is y=4x^3.
>>
>> Consider the standard AT cut which has turnover points at -15° C and
>> 65° C. The lower turnover would ordinarily not be used in ovens. A
>> set point error of ±1° C in the upper turnover point at 65° C results
>> in a frequency error of +14.875·10^-9. For cut 0, that same ±1° error
>> in room temperature results in a frequency error of ±31.25·10^-12.
>> This is an improvement of 476 to 1. You apparently have not thought
>> thru what improvements are possible with thermal coolers/heaters.
>> Among these is near instant warm up and greatly reduced power for
>> thermal management.
>>
>> πθ°μΩω±√·Γ
>> WB0KVV
>>
>> On Friday, June 2, 2017, Bob kb8tq <kb8tq at n1k.org> wrote:
>>>
>>> Hi
>>>
>>> Any real crystal you buy will have a tolerance on the angle. In the case of a crystal cut for turn
>>> the temperature will be a bit different and you will match your oven to it. If you attempt a zero
>>> angle cut, you will never really hit it and there is no way to compensate for the problem.
>>>
>>> Bob
>>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the time-nuts
mailing list