That's the next level of error detection. Unfortunately I notice with MSF
and these cheap receivers it can be difficult to determine what the binary
values being transmitted during the second were, which was my motivation
for the analysis I did.

I notice that there is one significant difference between DCF and MSF - the
former does not entirely turn off the carrier during the second, whereas
the latter does. I suspect that is the reason why with MSF receivers it can
so often be difficult to determine what the binary values transmitted were,
as the receiver AGC circuitry possibly has a difficult time keeping up with
drastically changing received power levels, which subsequently give rise to
noise in the received signal.

Splitting the MSF received signal into 100ms chunks, all of the seconds
apart from the start-of-minute marker are of the form 0AB11111. Using "x"
to represent a 100ms chunk whose value could not be determined, I notice
that many of the received seconds were of the form "0AB1x111" or "0AB11x11"
etc - i.e. there was only one 100ms chunk within the second whose value
could not be reliably determined and whose value was non-critical. With a
blame algorithm in place it should be possible to recover these signals.


On 6 February 2018 at 19:37, Poul-Henning Kamp <phk at phk.freebsd.dk> wrote:

> --------
> In message <CAGJ4F+6Nvic=RoUH7EozhKLCDziBTorMGu0iKj-sK6iXjw+zsQ at mail.
> gmail.com>
> , "Deirdre O'Byrne" writes:
>
> >I've been trying to see if I could design a decoding algorithm that
> >would be more noise-tolerant than the algorithms I've seen out in
> >the wild.
>
> You can: I baptised it "the blame algoritm".
>
> The trick is not to try to accept pulses as valid but to try to
> throw out pulses which are impossible.
>
> Imagine you have a 120 second long shift-register, and you feed
> your received pulses into it.
>
> Then try brute force, for every one of the newest 60 positions if
> that can be the start of a minute or not, by testing all the
> constraints you can think of, and there are surprisingly many.
>
> Some are obvious, the bits encoding the hour cannot contain "39",
> but that is a remarkable weak filter that seldom kicks in.
>
> A much stronger filter is that the bits encoding the hour must be
> the same as in the previous minute *unless* minutes were 59 in the
> previous minute *and* zero in this minute.
>
> If you count it up, that is a strong and very peculiar relationship
> on all the hour-bits and all the minutes-bits and if even one single
> of them are wrong, you can definitively discard that theory for the
> start of the minute.
>
> A similar thing holds for the date bits, the time in the
> previous minute must be 23:59:59 and in this 00:00:00 for
> there to be any difference between the dates, and even
> then, only a small number of possible changes in the date
> bits are valid.
>
> If you look in http://phk.freebsd.dk/phkrel/NTPns.20080902.tgz
> you will find a file called dcf77_blame.c with my code,
> here is a couple of the simpler tests:
>
>         /* LSB of minutes must be different from previous minute */
>         j = ip->shiftprev[(offset + 21) % 60];
>         if (j * ip->shiftreg[(offset + 21) % 60] > 0)
>                 FAIL((why, " 0"));
>
>         /*
>          * If the LSB of minutes was '1' in previous minute
>          * the next higher bit must have changed, if it was
>          * a '0' it must not.
>          */
>         if (j *
>             ip->shiftreg[(offset + 22) % 60] *
>             ip->shiftprev[(offset + 22) % 60] > 0)
>                 FAIL((why, " 1"));
>
> When using this algorithm, missing pulses is almost a
> non-issue up to around 40% of them missing, and even
> in an enviroment like that, it is not uncommon to see
> the algorithm lock on to the minute in about 34 seconds
> and know the full time in less than 3 minutes.
>
> If you make your pulse width discriminator *really* selective,
> which you might as well, you can "blacklist" disproved
> minute positions for the next many minutes as the
> risk of a '1' and '0' being confused is close to zero.
>
> That will get you minute lock, even with 70%-90% missing
> pulses, in a matter of minutes if you use a longer
> shift register.
>
> I did a parallel prototype for MSF, but I didn't need it
> so I never completed it, not sure I have it around any
> more.
>
> I should really write an article about that code...
>
> Poul-Henning
>
> --
> Poul-Henning Kamp       | UNIX since Zilog Zeus 3.20
> phk at FreeBSD.ORG         | TCP/IP since RFC 956
> FreeBSD committer       | BSD since 4.3-tahoe
> Never attribute to malice what can adequately be explained by incompetence.
>


More information about the time-nuts mailing list