[time-nuts] Phase Noise Comparisons
Bob Bob
bobbob55 at hotmail.com
Sun Jan 17 03:37:08 UTC 2016
Actually, I just acquired a HP3048a phase measurement system and I’m measuring all of the my signal generators, one of which is an 8660c with the 86603 plugin. Attached (ignore the caption) is the result compared against the 3048a internal 10mhz low noise reference.
David
> From: rob at nc0b.com
> To: time-nuts at febo.com
> Date: Sat, 16 Jan 2016 20:05:18 +0000
> Subject: [time-nuts] Phase Noise Comparisons
>
> I did some interpolation to compare the specs for phase noise of the 8660 the way it was done then to the present method. I think it is going to be something like -105 dBc/Hz to -110 dBc/Hz at 10 kHz offset at 14 MHz carrier frequency.
>
>
>
> As a comparison, the 8662A is about -135 dBc/Hz and the 8642A is better than -150 dBc/Hz.
>
> Actual measurements on multiple generators in my lab. Unfortunately I have never used or measured an 8660.
>
>
>
> 3335A -128 dBc/Hz
>
> 3336C -120 dBc/Hz
>
> 3325A -115 dBc/Hz (original version. Later improved by 4 pr 5 dB after the 3336C came out.)
>
>
>
> As a comparison a Rigol DG4162 is -115 dBc/Hz All at 10 kHz offset and on the 20 meter band.
>
>
>
> It depends what you are trying to measure. Sensitivity or noise floor, anything will work within the range of specified output accuracy and leakage. On the other hand, if you are trying to measure the dynamic range of a receiver, or its phase noise characteristics, you cannot do that with most generators.
>
>
>
> Rob, NC0B
>
>
>
>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 8660c 10mhz.png
Type: image/png
Size: 46659 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20160116/fd0dca59/attachment.png>
More information about the Time-nuts_lists.febo.com
mailing list