[time-nuts] Re: HP / Agilent 53132A
    Bob Camp 
    kb8tq at n1k.org
       
    Tue May  2 02:49:43 UTC 2023
    
    
  
Hi
Some, but not all, of the counters we saw die also had fan issues. I agree that once
the fan goes, the power supply probably is running on borrowed time. Cleaning the 
fans and checking their function was part of the yearly âcalibrationâ of the counters. 
I have no idea how may fans got replaced as part of that process. 
Bob
> On May 1, 2023, at 9:13 AM, Richard (Rick) Karlquist via time-nuts <time-nuts at lists.febo.com> wrote:
> 
> Not to argue with you, but it seems inconceivable that any fan would
> last 20 years, going 24/7.  Maybe the fans quit, and the instruments
> continued to run w/o them.  Might be OK if in a cool room.
> 
> 73
> Rick N6RK
> 
> On 4/30/2023 2:51 PM, Michael Wouters via time-nuts wrote:
>> Curious. We had about eight 5313x counters operating until a few years ago.
>> These were in logging systems and cal setups, where they were never powered
>> off, some for 20+ years. There were no power supply failures.
>> Cheers
>> Michael
> _______________________________________________
> time-nuts mailing list -- time-nuts at lists.febo.com
> To unsubscribe send an email to time-nuts-leave at lists.febo.com
    
    
More information about the Time-nuts_lists.febo.com
mailing list