Hi Fred,
there are three things to consider:
1) alignment of the 1PPS from your GNSS simulator to the simulatorâs RF output. GNSS simulators are typically aligned to L1 C/A signal with different level of accuracy.
2) 1PPS cable delays, and 1PPS measurement threshold - those can influence your absolute 1PPS offset from GNSS simulator.
3) ZED-F9T module carrier board. I am assuming you have ZED-F9T module soldered on something. Even though ZED-F9T may have been calibrated in production, as soon as it undergoes solder heating cycle, this can easily shift the group delay of internal components of ZED-F9T.
I think best way to go around this is to calibrate 1PPS output against good GNSS simulator after the ZED-F9T is soldered on the PCB. No more solder heat cycles after calibration!
Best regards
Zdenek
> On 29 Apr 2024, at 11:36, rf lab via time-nuts <time-nuts at lists.febo.com> wrote:
>
> Hi All,
>
> I have a ublox ZED-F9T with a very short RF trace to an SMA connector (<
> 1cm), so I expect the delay due to the RF line to be in the ~100's pico sec
> range. Using a GNSS simulator I measure the average 1pps delay relative to
> the simulator to be about 25ns. This figure is quite close to a typical
> GNSS SAW group delay.
>
> My question is if anyone has done something similar or knows what the
> actual delay is for a ZED-F9T module.
>
> Thanks
> Fred