Sali Tobias,
On Thu, 17 Jul 2025 11:51:39 +0200
"Pluess, Tobias via time-nuts" <time-nuts at lists.febo.com> wrote:
> I am in the process of designing a new GPSDO. As my previous one worked
> very nice, but had several small design issues, I want to make a new one
> that is a bit improved. So I have drawn a new schematic that is in the
> attachment.
Would you mind sharing what kind of issues you had?
Your design looked very nice, as far as I remember.
> I use again the Oscilloquartz / "UTC" branded oscillator and an integrated
> dual buffer / distributor and a TDC7200 as interpolator. I did not find any
> other time to digital converter that has both, higher resolution as well as
> a package that can be hand soldered, so I stick to the 7200. I am happy,
> however, if someone can give me a hint for a possibly even better TDC.
I don't think you'll get much better results using a different TDC.
The noise in the sawtooth corrected PPS is still several orders of
magnitude larger than resolution of the TDC7200. The only important
bit here is that the uncertainty contribution (noise, bias, etc)
of the TDC is below that of the PPS output of the GPS, which holds
for the TDC7200, as far as I am aware of.
> Also, with this new GPSDO, I want to add a serial connector that connects
> also the 1PPS signal, such that it could be used by a PC with GPS/NTP
> software, such as chrony. I was also thinking about PTP, but I found no
> elegant solution for this.
The canonical way is to use pin 1 (DCD). and then use the PPS subsystem
of your OS to get accurate timestamping. All ntp clients I am aware of
support this.
> Also, I would like to increase the resolution of the DAC that I use to
> steer the OCXO. For this, I have 2 optional solutions. One is the PWM DAC,
> that I copied from the Oscilloquartz STAR4+ GPSDO I have. The other one is
> the classic 16-bit DAC, and here, I would like to use later some kind of
> delta-sigma modulation to increase the resolution. For this reason, there
> is this low pass filter after the DAC.
I am pretty sure that's not just a PWM but a delta-sigma modulator.
I would go with the same on the DAC. Implement a second oder delta-
sigma modulator on your µC and filter your output. This would give
you log_2(sampling rate / BW) * 2.5bits additionally to the 16bit
you have from the DAC8551. I.e. If you sample at 100ksps and have
an output filter bandwidth of 100Hz, you'll get log_2(100k/100)*2.5bits
which is about 24 bits. Realistic would be some 5-15 bits more, depending
on the exact behaviour of the DAC. Please note that this approach
only works because the non-linearity due to the DAC's behaviour
will be compensated by the control loop. We can't easilly build
DACs with an ENOB of more than 20bits (there are ways, but they
become expensive very quickly and have lots of caveats)
>
> Also I would like to add something new to this, if the computing power of
> the microcontroller is sufficient. I would like to collect some statistical
> data about the OCXO aging, store this data in the flash memory that I added
> for this purpose, and then try to extrapolate the aging. Then, in the case
> of holdover, use the extrapolated aging data to steer the OCXO such that
> its aging is compensated. Not sure if this is even possible, but I would
> like to try. With my current design of the GPSDO, this is not possible, as
> there is not enough storage for some statistical data.
That's a very nice idea, but I think your major limitation will be
the amount of data you can store, if you want to do long term analysis.
If your goal is just to estimate the aging, than that's a much simpler
problem. But you will need to remove all confounding factors first.
At the very least, you need to estimate your linear temperature dependence.
I doubt that the quadratic component of the temperature dependence is
significant for the 8663 clones, though I have not measured them.
An easy way to do this, both estimating temperature dependence and
aging, is to use a Kalman filter. This will give you a decent estimate
with low memory and computational effort. Start first with a linear
aging estimator and, once you got it stable, use a quadratic one.
> What do you guys think about my design? Would it be a nice GPSDO or
> complete crap ideas.
The ideas are very nice and I wholeheartedly support you.
Some small nitpicks on the schematic:
The buffer inverter used for the 10MHz clock is biased by a
resistive divider. Use a 100k-1M resistor feeding back from the
output to the input instead, to self-bias the inverter. This way
you eliminate changes in the threshold voltage of the inverter
due to temperature changes, which would give you a slow drift
in the apparent delay through the inverter.
You are using quite a few ferrite beads in the power path of
components. Be careful with them. for low frequencies (below 1-10MHz)
they act purely inductive with almost no resistive component.
I.e. you get a nice, high-Q resonator with the capacitors. Make
sure your design does not accidentally excite any of the resonances
in your system that these ferrite beads introduce. Better use point-of-load
regulators whenever you need low noise or high PSRR for senstive components.
> Also, I was thinking about adding a 100MHz output. But I am not sure if
> this really adds value to it, as I see that 100MHz are used by some
> instruments, but only rarely, and I am not yet aware of good, obtainable
> 100MHz oscillators.
We kind of standardized on 10MHz for most things, so having a 100MHz output,
if you don't have a need for it, does not add much benefit. At least, I don't
know of many instruments that use 100MHz reference inputs. If you need to
have a frequency locked 100MHz, I would do that as an additonal board.
This way you can optimize it for whatever you need it for.
I hope this helps.
Es schöns Wuchenend no und Gruess