[time-nuts] upgrading agilent 53132 counter
Bob Camp
lists at cq.nu
Fri Oct 2 19:21:27 UTC 2009
Hi
One easy way to get the "eprom fail" message on the 53132 is to write
settings to memory "a bit to often". I blew out a set of them in under a
year back when they first came out...
They still worked fine, but they did come up with the error at boot. I
suspect that they also did not quite save settings correctly anymore.
Bob
-----Original Message-----
From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On
Behalf Of David
Sent: Friday, October 02, 2009 3:03 PM
To: time-nuts at febo.com
Subject: Re: [time-nuts] upgrading agilent 53132 counter
Hi Alain
I can't help with the differences in the PSUs, there is little you can tell
by looking at the unit. I wonder if it might be easier to run the OCXO from
a new additional supply housed in the unit, there appears to be space for
this (?) and feed the 10MHz into the external REF I/P? There are options.
> effectively, at power on, i have a message : rom failed, but
> it seems to have no importance, since after actioning the
> run touch, it works well (why ???). Do you really need an
> image of the eprom ?
Regarding the eproms, your fault description matches what I've seen in 3 or
4 units, all at first glance functional. The worry is that the fault clearly
was not there a few years ago and you have got to assume the worst, why
should it stop degrading? The blank parts are Am29F010B flash and were cheap
and available when I looked. Let me know if you want the image files I
generated of good parts emailed, the set of images etc in various formats is
about 1.7M (4 parts in 3 different file formats).
David
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
More information about the time-nuts
mailing list