[time-nuts] Re: DMTD Questions
Chris Hastreiter
chastreiter at gmail.com
Sat Jul 11 14:32:55 UTC 2026
Alex,
The "additive" measurement method you're describing sounds exactly like
what I used to measure the residual noise of my synthesizers. In the case
of my 8663A I used its internal reference but measured the phase noise of
its output referenced to the "ref out" connector on the back. In the case
of my 33250A I used an external 10811A as the reference. Here are my
results, it looks like the 33522B is on a similar order of magnitude. (the
8663A is the lower noise trace on my plot)
I've also included the absolute noise measurement plots for comparison.
Chris
On Thu, Jul 9, 2026 at 4:50â¯PM Alex Denner <alex.denner at gmail.com> wrote:
> By coincidence, I happen to be looking at one small part of your
> project. I was curious about using a DDS to do frequency offsetting
> for another purpose. In these graphs, I used the Keysight 5047A (the
> AnaPico PN analyzer) cross-correlation phase noise analyzer. I looked
> at the phase noise of a free running Keysight 33522B vs the same unit
> with an external reference. Below are the two phase noise curves from
> 0.1 Hz to 5 MHz (10 MHz carrier). One can see the free running 33522B
> adds a decent amount of jitter if using it for a synthesized offset
> generator. I was estimating jitter by the difference in the two
> integrals. (I think the graphs have 100 correlations with some spur
> removal and averaging, I could try to find the details if anyone
> cares.).
>
> However, I thought of an interesting experiment: I could run an
> "additive" phase noise plot of the 33522B where one channel is the
> external reference and the second channel is the 33522B's (locked to
> external reference) output where the output is set to be in quadrature
> to the external reference entering the 5047A. It is not really an
> additive phase noise measurement in the classic sense but it may
> measure the total phase noise contribution of the DDS synthesis chain.
>
> As you know, the unreferenced HP 8663a with the OCXO is quite a bit
> better---probably some combination of the OCXO, better synthesis
> chain, and better power supplies, etc. I will also include a phase
> noise plot of my 8663a from 0.1 Hz to 5 Mhz at a 10 MHz carrier.
>
> Alex
>
>
> On Thu, Jul 9, 2026 at 4:42â¯AM Chris Hastreiter via time-nuts
> <time-nuts at lists.febo.com> wrote:
> >
> > Time-nuts,
> >
> > I have been working on some modifications to Corbyâs DMTD design [1] in
> an
> > effort to further improve its noise floor and had a couple questions for
> > the group. The performance Iâve been able to achieve so far is shown in
> > the below plot with the blue trace being the noise of the ProTIC counter
> > itself, the pink trace being the combined noise of the counter and my
> zero
> > crossing detector (both channels driven by a 10Hz sine wave), and the
> > remaining traces the full system noise floors with 4 different offset
> > oscillator options. All noise floor measurements were taken with a
> carrier
> > frequency of 10MHz and an offset of 10Hz. I believe that much of the
> noise
> > floor degradation at long tau is due to thermal effects in the first
> stage
> > amplifier of the ZCD as I havenât yet gotten the system installed into an
> > enclosure and I can worsen the issue by thermally disturbing the LT1028
> of
> > the first stage.
> >
> > [image: image.png]
> >
> > My first question is in relation to mixer noise which is a noise source
> > that I have not yet considered. I am using the same SBL-1 mixer as
> Corbyâs
> > original design but [2] implies that some mixers exhibit lower flicker
> > phase modulation than others and I havenât seen any literature or
> > specifications that point to any devices other than the 10514A in that
> > paper. They also discuss adding capacitive loading to the mixer to
> further
> > reduce noise referencing another paper that I have yet to be able to
> find.
> > Is there any expectation that changing to a mixer different than the
> SBL-1
> > will improve my results and should I consider capacitive loading on the
> > mixer? If so, how would I determine an appropriate amount of capacitive
> > loading? I currently have it loaded with a bridged-T filter holding 50
> > Ohms of impedance across the frequency spectrum.
> >
> >
> >
> > My second question is in regard to the common oscillator noise
> > contribution, specifically for the use case of a synthesized offset
> > oscillator. The below plot shows the results Iâve gotten so far with
> some
> > sources that I already have, and as can be seen the system noise floor
> > degrades much faster with the synthesized offset than with the 10811A.
> > However, it would be advantageous for me to be able to use a synthesized
> > offset for tests where the DUTs differ slightly in frequency and the
> system
> > will phase wrap. The reason for this is that when using a free-running
> > oscillator the phase wraps are very difficult to remove automatically
> > because the divided phase wrap will differ from the expected with any
> > frequency error in the offset oscillator. Do you have any suggestions
> for
> > synthesizers or other offset techniques that would help with this?
> Using a
> > 10811A and a slow PLL as a cleanup loop on the synthesizer does appear to
> > work well in the below plot, but looking at the first plot in this email
> at
> > least in my initial implementation there seems to be some low frequency
> > oscillation in the PLL impacting the noise floor. Iâve also included a
> > plot showing measurements Iâve taken of the residual noise for the two
> > synthesizers used as well as a comparison of two of the 10811A
> oscillators
> > that Iâm using.
> >
> > [image: image.png]
> >
> > [image: image.png]
> >
> > [1] https://www.eevblog.com/forum/metrology/dmtd-board/
> >
> > [2] https://ieeexplore.ieee.org/document/5075023
> >
> >
> >
> > Thanks,
> >
> >
> >
> > Chris
> > _______________________________________________
> > time-nuts mailing list -- time-nuts at lists.febo.com
> > To unsubscribe send an email to time-nuts-leave at lists.febo.com
>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 33250Avs8663AResiduals.PNG
Type: image/png
Size: 145096 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20260711/d12c693d/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 33250Avs8663A.PNG
Type: image/png
Size: 133616 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20260711/d12c693d/attachment-0001.png>
More information about the Time-nuts_lists.febo.com
mailing list